I sense some confusion. If you
* just want to "import" the volt to your lab, then why not participate in one of the existing "calibration clubs"?
* build/sell a device for such "calibration clubs" (tiny market), such a device would be a "transfer standard" and as such needs to be
-- low noise to allow determining the DC value with great confidence
-- ruggedized (as shippers and members *will* drop them, *cough*, *cough*)
-- protected input (overvoltage, reversed polarity) and output (short circuit)
-- small and light to reduce shipping costs
-- withstand temperature extremes while showing only minimal hysteresis
-- should use a standard voltage (so that it can be compared easily), that'll be 10V these days and
-- standard connectors 4mm "banana" 5 way binding posts pretty please with possibility to screw in bare copper wires
-- not too expensive, as it might get lost during shipping
it does *not* need to be
-- long term stable, as it is "re-calibrated" (or current value recorded) every few days (few weeks at the most), when returning to home base
* If, on the other hand, you want build/sell a home reference (much larger market, but apparently not large enough for e.g. Mr. Geller's reference to be economically viable), you want it to be first and foremost long term stable (which is time consuming and hence expensive to determine and demonstrate)
It should not require batteries during shipping, which entails a whole bunch of problems. In today's world, things that show batteries inside on an X-Ray, will get flagged for further investigation-- and may get lost or destroyed as a consequence of that. It should not be affected by the input voltage much (with an input range of say, 12V to 16V)-- and it should not be affected by humidity or barometric pressure.
What this means is that every reference that is NOT in a hermetically sealed package is NOT acceptable. Unfortunately, almost all hermetically sealed references are being phased out by the idiotic bean counters at the various semiconductor companies, with the possible exception of the LM399 and the LTZ1000.
The (heated) LM399 has lower mechanical hysteresis than even the LTZ1000A, let alone the LTZ1000. The LTZ1000(A) references are better off if they are continuously powered.
So, this absolutely SCREAMS to use the LM399. You *could* use 4 of them in parallel to get the low frequency noise down, without getting too ridiculous on the final price of the reference. I once talked to my Linear Tech rep about the LM399. I asked what the big DMM manufacturers were doing to sort these into acceptable parts. He would not say what they were doing, but he did tell me that only about 1 out of 10 references ever made it into the 6-1/2+ digit meters. They are selecting for low 1/f noise (1Hz and below), low popcorn noise, low temperature drift, and low temporal drift. Some of these have to be monitored for over a year to find the "shining jewels". Probably, they will need some kind of artificial aging schedule before the drift testing is done. This could easily turn a $7 part into a $70 part (plus labor and fixturing).
Now, the LM399 has very poor initial accuracy, but very good stability. There are some very old DMMs that have this reference in them, that have a surprisingly low annual drift rate (less than 1ppm), but this is only after 10 years or so of power-on time. Using the LM399 would mean that you need to have a highly adjustable 7V to 10V boost circuit that does not drift over time, and this (in turn) indicates a PWM scheme. This also provides the opportunity to trim the reference for temperature and barometric pressure related drift. There is already work being done on this on the LM399 thread here:
https://www.eevblog.com/forum/metrology/lm399-based-10-v-reference/This could be initially adjusted and calibrated at a Cal-Lab (ENI Labs maybe?)-- and then calibrated each year just after the Cal-Lab gets their 732B calibrated at Fluke for minimum uncertainty. (Here, "calibrated" means "accurately measure" and "adjust" means "change the absolute output value closer to 10V".)
This reference could then be circulated around through the Cal-Club, so users could calibrate and/or adjust their in-house 10V standards, DMMs, and calibrators.