Author Topic: Leeds and Northrup 86962 millivolt meter  (Read 1313 times)

0 Members and 1 Guest are viewing this topic.

Offline Wallace GasiewiczTopic starter

  • Super Contributor
  • ***
  • Posts: 1339
  • Country: us
Leeds and Northrup 86962 millivolt meter
« on: April 07, 2022, 01:02:40 pm »
Leeds and Northrup 86962 millivolt meter
Just picked up one of these as a lark. Looks pretty good inside and out except the  batteries are corroded and battery terminals need some care.
I have trouble finding a manual. This thing has a Weston Cell in it. Ca I measure the cell with an HP 3456??

Wally
 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2903
  • Country: 00
Re: Leeds and Northrup 86962 millivolt meter
« Reply #1 on: April 07, 2022, 01:38:56 pm »
The 3456A should have an input impedance of > 10 GOhm in DCV ranges up to 10V, so that should be fine to measure a Weston cell if you set it to manual ranging just to be sure it doesn't switch to the 100V range.
 