Here is an example using the FX reference and the ratio function of the 34401A to track the tempco of a Fluke 731B 10V reference. I knew the tempco wasn't great on this unit, I had roughly estimated it to be 1.5 ppm/F (ambient) from previous usage and testing. The line, which I just drew manually, represents -2.5ppm/C. Note that here I'm measuring the temp of the internal case around the ref amp and precision resistors rather then ambient. I'm using a K-type bead thermocouple with Kapton tape to hold it to the case, reading with a Fluke 289. This leads to some interesting issues that anyone mixing temperature into their logging needs to account for--look at the the graph and see if you can spot the anomaly.
Anyway, given the outstanding performance of the two meters in question there isn't too much improvement to be had, but there is a little. You can see the 34401A very closely tracking the 8846A with a delta of between 4 and 5ppm, which would be approximately the sum of the nominal difference (from 10V) of the FX reference plus the previously measured error of the 8846A (about 0.7ppm). If I had a worse meter with a DC ratio function I'd use that for an example, but I haven't characterized my two monsters, the 8505A and 8506A, quite yet. And I'm not even sure they're "worse".