Thermal voltages?
Could be, though I tried to avoid that. Photo shows test setup. Everything had been on for two weeks before the data was collected. Data showed that meters stabilized in about four hours. All identical test leads to breakout panel. Breakout panel has BNCs wired to copper PWB with identical materials and similar lengths to pickoff point for cabling leading to 3457 in. Those wires are all identical materials and similar lengths, soldered to breakout board and tied to standard 44492 screw terminals on back of 3457. Room temperature was relatively stable less than 1 deg C variation, and uncorrelated with observed short term drifts seen in all channels.
Another possibility is some sort of noise synchronous with the sampling, since the channels are collected serially, not at the same time. Haven't been able to come up with anything of this sort either.
The EZGPIP script used is included for reference. Renamed to text file to allow posting to forum. If used in EZGPIB just rename back to .488 file.
Still plenty to think about before the reference arrives. Fortunately there is lots of time on the calendar, who knows how much will be available for this project.