Hi TiN & Andreas,
>>> I have to ask though, why 2 NPLC ? If I read scales correctly, your source tempco is around 1ppm/C, which is bad unless you are not measuring LTZ output? I'd think NPLC 10 or 50 to be used for such a test, to rule out ADC.
I chose 2 NPLC to be able to measure ~DC-10Hz noise as well as drift. I agree with you that it picks up more meter noise that way - ~0.04 of RMS noise and ~0.15ppm of 'gain error' @ 7V reading if I read the specs correctly. I guess those will average out though and it is interesting for me to see when the noise spikes are - I have not located the source yet.
>>> Also with 0.9ppm/day (similar to 1ppm/day I had on second A3 PCBA here) I'd consider A3 PCBA "broken". You can check if it's A9 LTZ drift or ADC by running ACAL. If ACAL brings readings back to initial value, your LTZ is stable, and it's A3 PCBA problem.
Yes it looks suspect and I'm going to leave the test running for a while to see what happens to #1045. It's still got v1 of the slave processor firmware so that may have something to do with it. I have ordered the 8051 (which is where I think the slave firmware is) from Keysight but they are being a bit random - yesterday they sent me a box 50x30x20cm with one button in it < 1x1x1cm, ho ho ho.
Many thanks for running that similar test. I'm away for the weekend but will run the data through the same analysis for comparison when I get back. BTW my simple test setup below - no screening atm. The cover for the ref btw is 3d printed, double walled, and on the bottom of the board also.
Andreas re:
>>> The LTC2057 has advantages at very low power supply voltages (more headroom). But I fear that it is more sensitive to EMI pulses (without the capacitors which I had to remove).
Thanks for that feedback, I also found, on another board, that I had to remove decoupling from input and output to stop it acting up. In particular, a 100nF on the output resulted in a nice 250kHz oscillation. I'm trying to remember what I once thought I knew re xfer functions and stability
![Smiley :)](https://www.eevblog.com/forum/Smileys/default/smiley.gif)
I was thinking of trying the LT1013, which is why I asked.
>>> Depending on which side of the LTZ I apply my hot air gun the output voltage rises or falls.
Well that's a fairly severe test - that probably happens (I have not tried it) with say the HP layout?
Alan
![](http://anagram.net/Nuts/MeterDrift/Images/IMG_2555.JPG)