As announced by Andreas we wanted to find out what is best: "Using a solid pcb or spend some slots around the heat controlled voltage reference?" and furthermore: "Keep the leads as long or as short as possible?".
So let's start the discussion ![Smiley :)](https://www.eevblog.com/forum/Smileys/default/smiley.gif)
Ok, I have finally evaluated the measurements = .csv tables of Branadic from the LM399 thermograpic measurement.
Long leads without slot: around 5 degres stray within pad + 1,7 degrees from pad to pad
Long leads with slots: around 5-6 degres stray within pad + 2,3 degrees from pad to pad
Short leads without slot: around 9-11 degres stray within pad + 3,1 degrees from pad to pad
short leads with slots: around 12-15 degres stray within pad + 6,6 degrees from pad to pad
So from the measurement the conclusion is that with long leads the stray (=thermoelectric voltage) within pad is lower than with short legs. And the thermoelectric voltage between different pads is lower without slot.
If there should be mechanical issues the slots should be at least around 15 mm from the reference to give the pads the possibility to equalize the temperature.
2 simple conclusions:
- short leads will increase the temperature of the PCB, long leads will give more thermal isolation between component and PCB => long leads preferred
- slots in the PCB give higher thermal isolation between solder joint area and outer PCB area, thereby reducing cooling of the solder joints through the PCB, and in turn increasing the local heating of the solder joints. Higher solder joint temperatures might cause temperature differences more easily and will increase thermo voltages.
I'm now very confident that 45°C and a PCB without slots is the most stable solution.
So the conclusions of Frank are now quantified.
Before the measurements I planned to make slots near the reference and short leads. Now I will go for long leads and at least keeping the slots far away from the reference. (e.g. at the edges of my inner metal shield).
@branadic
When regarding the videos (the flimmering of the heat) I had the idea that the result should be also be visible by noise measurements. So its a pity that we did not measure the noise voltages of the references in parallel to the thermographic measurements.
With best regards
Andreas
============================== evaluation in detail ==============================
pin upper left:
min 27,64
max 32,81
avg 29,59356083
stddev 1,144305683
max-min 5,17
pin upper right:
min 28,02
max 33,15
avg 30,25453333
stddev 1,000260225
max-min 5,13
pin lower left:
min 27,8
max 33,18
avg 29,33930796
stddev 1,031510593
max-min 5,38
pin lower right:
min 28,87
max 33,48
avg 31,08652597
stddev 0,855853162
max-min 4,61
all in all: around 5 degres stray within pad
pin upper left:
min 30,1
max 35,46
avg 32,78116883
stddev 1,173468631
max-min 5,36
pin upper right:
min 28,44
max 33,54
avg 30,48031128
stddev 1,148254067
max-min 5,1
pin lower left:
min 29,24
max 34,46
avg 31,80304
stddev 1,342351664
max-min 5,22
pin lower right:
min 29,3
max 35,7
avg 31,30611111
stddev 1,183186142
max-min 6,4
all in all: around 5-6 degres stray within pad
short leads without slot
pin upper left:
min 36,649
max 47,527
avg 42,58773745
stddev 3,146954606
max-min 10,878
pin upper right:
min 38,599
max 49,453
avg 43,66859498
stddev 2,401555908
max-min 10,854
pin lower left:
min 37,141
max 45,726
avg 41,45009881
stddev 2,115676062
max-min 8,585
pin lower right:
min 38,552
max 48,064
avg 44,64691221
stddev 2,080904881
max-min 9,512
all in all: around 9-11 degres stray within pad
3.1 degrees from pad to pad
pin upper left:
min 41,186
max 55,277
avg 46,75962041
stddev 3,418151728
max-min 14,091
pin upper right:
min 39,226
max 50,78
avg 44,88190295
stddev 3,051374117
max-min 11,554
pin lower left:
min 42,81
max 58,327
avg 50,25318644
stddev 3,742872955
max-min 15,517
pin lower right:
min 40,915
max 56,608
avg 48,35819758
stddev 4,061728375
max-min 15,693
all in all: around 12-15 degres stray within pad