Author Topic: Ultra Precision Reference LTZ1000  (Read 1415141 times)
Offline Andreas
Re: Ultra Precision Reference LTZ1000
« Reply #2475 on: January 30, 2019, 08:31:16 pm »
Interesting.
Pardon my ignorance, but how did you come up with the TC resistor values? Purely empirical?
I have a non-A LTZ1000 and thinking what values I should start from.
Hello,
I would start with R9 = infinite.
you can trim T.C. up by +40 ppb/K for each 1 Meg resistor in parallel for R9 on a LTZ1000A.
https://www.eevblog.com/forum/metrology/ultra-precision-reference-ltz1000/msg907048/#msg907048
Hello
my heater is an old ion. (Bügeleisen)
Do not take it too serious.
Just a fast solution for me. :-)
Interesting trimming bolt on the side  :-DD
best regards
Andreas
The following users thanked this post: alex-sh, ha110ween
Online Kleinstein
« Reply #2476 on: January 30, 2019, 09:15:08 pm »
The TC compensation used the heater voltage. This means the strength depends on the used power level and set temperature. So the +40 ppb/K is only a ball part number. The actual number depends on the heater level and thus set temperature and insulation (e.g. cap).  A lower set temperature is expected to give less heater power and thus more change in voltage to compensate for temperature changes at a given thermal setup. The tendency should be a larger resistor for a lower set temperature.
Rather close to the set temperature things can get quite nonlinear - making TC compensation difficult.
Testing / selecting R9 should be done with the final thermal setup (especially cap over LTZ) and set temperature.
The following users thanked this post: TiN, alex-sh
Offline alex-sh
« Reply #2477 on: January 31, 2019, 01:55:35 pm »
Hello,
Thank you
The below is very helpful:
For R9 = 1 Meg I get a T.C. increasement of around +0.04ppm/K
100K give around +0.4ppm/K over 20 deg C.
I feel like to need to read all 100 pages of this thread.  |O
Offline Andreas
« Reply #2478 on: January 31, 2019, 07:36:28 pm »
Otherwise you do not know that shortening the legs of a LTZ1000A also increases the T.C.
I recommend to start with long legs and shorten as necessary.
with best regards
Andreas
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #2479 on: February 01, 2019, 12:39:59 am »
Hmm, you have a link to data on this? I didn't see a difference in my boards.
Offline Andreas
The following users thanked this post: DC1MC
Offline DC1MC
« Reply #2481 on: February 01, 2019, 09:17:36 pm »
Woah, this legs thing it's really woodoo magic  :o !!!
 DC1MC
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #2482 on: February 01, 2019, 10:13:10 pm »
Andreas
I'm lost with your posts now, as you contradicting yourself in here  :-//.
Offline Andreas
« Reply #2483 on: February 02, 2019, 06:12:42 am »
Hello,
sorry I do not see what you mean.
I think it is a straight forward observation:
the shorter the legs the more less negative the T.C. (so T.C. increases).
with best regards
Andreas
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #2484 on: February 02, 2019, 12:00:06 pm »
Per your results, going from -80 ppb/K to -22 ppb/K and from -230 ppb/K to -95 ppm/K means TC is decreased (less impact on output voltage from temperature), not increased.
Online 2N3055
  • Country: hr
« Reply #2485 on: February 02, 2019, 12:05:45 pm »
It decreased in absolute value (mathematically) , but it increased in value (as in it was less negative)....
Terminology gap.
EDIT: Spellchecker typo. Value decreased, it didn't die ...  :-DD
« Last Edit: February 02, 2019, 02:08:34 pm by 2N3055 »
The following users thanked this post: Andreas
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #2486 on: February 02, 2019, 12:29:03 pm »
Yup. I'm going by usual convention of zero TC = no change, ignoring the sign.
Offline GregDunn
« Reply #2487 on: February 05, 2019, 01:11:47 am »
Freshly arrived from across the pond, thanks to branadic for this nice little package.  Aged LTZ1000 with log of voltage vs. time, measured at 10.000002 VDC.
Obviously, I don't have a DMM good enough to check the LSDs; but this newly repaired HP3490A seems to think it's pretty close at 5.5 digits worth of precision.  This is after 4 hours of warmup; I'm now in the process of letting some other DMMs settle so I can cross-check them with it.
Offline AG7CK
  • Country: th
« Reply #2488 on: February 05, 2019, 02:02:48 pm »
I like this "no voodoo" clean simple functional 10V box based on the green PCB from recent posts and a "classic" DIY-electronics style enclosure.
If Branadic or someone else has one more for sale or will produce these characterized plug-and-play reference units for other members, I would also like to buy one or be put in a queue for future purchase. Thanks.
Offline alex-sh
« Reply #2489 on: February 07, 2019, 12:33:54 pm »
My board is almost done. Still waiting for precision resistors to arrive...
Offline MiDi
  • Country: ua
« Reply #2490 on: February 11, 2019, 01:09:40 am »
My #1 branadic / Dr. Frank
Did quick check on noise for 7V & 10V with linear bench PSU:
Noise floor 10min: ~115nVpp
7V buffered
10min: 2.3µVpp
12s best: 1.5µVpp
12s typ: 1.7µVpp
12s worst: 2.1µVpp
10V
10min: 3.2µVpp
12s best: 2.0µVpp
12s typ: 2.3µVpp
12s worst: 2.8µVpp
« Last Edit: February 18, 2019, 12:14:08 am by MiDi »
Offline Magnificent Bastard
  • Country: aq
« Reply #2491 on: February 11, 2019, 02:08:11 am »
You might get even better noise readings if you add an air-draft excluder to the top and bottom of the LTZ1000.
Offline RandallMcRee
« Reply #2492 on: February 11, 2019, 04:55:32 pm »
I suspect that you have some unidentified noise source.
Both of my DrFrank boards (sorry for not posting!) showed ~0.9uV of noise. There were two different LTZs, one from Arrow, one from Mouser, A and non-A.
Certainly air drafts, etc. are a possibility. My noise floor is similar to yours.
Offline MiDi
  • Country: ua
« Reply #2493 on: February 11, 2019, 11:37:30 pm »
Thanks for the suggestions, seems it is worth to investigate.  :-+
First thought was: ok, it is in spec, but bit on high side (DS: 1.2µVpp typ, 2µVpp max).
Online Dr. Frank
« Reply #2494 on: February 12, 2019, 03:50:11 pm »
I'm very pleased that Mr. Pettis' resistors fit well on the PCB, that looks really nice, also in the final enclosure.
I also suggest more thermal isolation, as that noise level is a bit high, but dependent on the sample.
That measurement is also dependent on the bandwidth under observation, so identical methods and BW limits should be obeyed.
For a more practical use, you may determine StDev with your 'new' 3458A, at NPLC 100, so you may get 100..200nV rms only, that's equivalent to maybe < 1Hz BW.
And you might measure and trim its T.C. 8) (if not already done?)
Frank
« Last Edit: February 12, 2019, 04:03:49 pm by Dr. Frank »
Offline MiDi
  • Country: ua
« Reply #2495 on: February 18, 2019, 01:02:05 am »
Seems it was a misunderstanding what was measured (clarified in the post):
I measured 7V after buffer and the noise added by the LTC1052 is quite high (DS: 1.5µVpp typ, 7Vdirect -> 7Vbuf: ~0.7µVpp), did not realized that it adds so much noise...
Battery powered 0.1-10Hz noise:
7V buffered: ~1.7µVpp (12s, typ) - same with linear bench psu
7V direct: ~1µVpp (12s, typ)
Seems it is in the ballpark of other users  :-DMM
There was no measurable difference in noise with or without air-draft excluder in 0.1-10Hz range for this unit, but maybe it affects noise below 0.1Hz  :-//
TC compensation leads to next project: thermal chamber
Or even two for direct comparison of temperature stabilized one against DUT?
Offline Insatman
  • Country: ph
« Reply #2496 on: February 19, 2019, 10:19:26 am »
Your results are generally consistent with mine.  Typical 0.1Hz to 10Hz noise is around 0.8 to 1.4uV with a noise floor of the 10K:1 amp used around 150nV.
Retired Pulsed Power Engineer/Physicist...now I just dabble in electronics
Offline GregDunn
« Reply #2497 on: February 28, 2019, 06:40:00 pm »
I've been using my reference to check calibration of my DMMs (particularly newly repaired ones) and as such I have had to move the output cables around to connect to different units.
Something I've seen happen: when removing and re-connecting cables, sometimes the reference output will drop to about 2.9VDC and stick there.  Cycling power to the reference brings it back immediately.  I've been careful to ground myself before touching the reference's case and not touch the bare leads but it happened anyway.
Is this a characteristic of the LTC1052 going into protection mode?  The datasheet only talks about overload recovery and doesn't address any sort of latch-up considerations...
Offline branadic
  • Sounds like noise
« Reply #2498 on: March 01, 2019, 10:37:58 pm »
Haven't observed something similar yet, neither on your reference nor on mine.
-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.
Offline chuckb
« Reply #2499 on: March 01, 2019, 10:57:28 pm »
From the LTC1052 data sheet.
Note 2: Connecting any terminal to voltages greater than V+, or less than
V–, may cause destructive latch-up. It is recommended that no sources
operating from external supplies be applied prior to power-up of the
LTC1052/LTC7652.
It usually takes several ma to induce SCR latchup. Is there a transorb on the output to protect against ESD causing latchup? Or it may just be a bad Op-amp.