Ah, many thanks for your clear description of optimum operating points, voltage and current noise vs. source impedance.
>>> I am just a little skeptical in expecting leading etch performance for a part that does not even have 1/f noise specs in the data-sheet.
Yes I understand and it seems odd that THAT corp is seemingly too lazy to document and test the noise specs of something they market for 'low noise front ends'.
I had guessed that, for AN159, that Linear must have measured the noise of a number of a BJTs and selected the best for their application. Presumably, not more that a day or so of work for a junior engineer if you have the right equipment sitting on the bench. If that was the case, I was simply piggy-backing on their work.
However, I noticed a couple of days ago (reading AoE on a flight
) that tables 8.1a and 8.1b have H&H's own BJT noise measurements. For multiple parts (table 8.1b), the THAT300 is shown as 0.9nV/sqrt(Hz) at 1mA i.e. well towards the good end of the table. So, Linear
might have duplicated H&H's work or they might have simply read the numbers in the AoE table.
Alan