Partial quote.
By the way: the resistors R14 and R15 are 1K as far as i have seen. 200R and 20K as in 845A_324_RevB_2-1.pdf don't make a lot of sense.
Dieter, after studying TiN's photographs again, I think I have to agree. R14 and R15 are marked 10011 and 11001 respectively, with one simply facing backwards. Not 20022 and 22002. If those stripes are brown not red, 1K makes more sense than 110 ohms, the alternate interpretation.
In that case, my explanation of the offset trim is bunkum. I don't know why R14 and R15 exist, unless it's to swamp out U4 and U5's small but variable on-resistance.
Fluke took the trouble to install them, along with their teflon terminals - it must have been worth it.
And if R14 == R15, then U4 and U5 must be selected so their mostly-canceled offsets result in the correct net polarity for R7 to trim back to zero.
DISCREDITED PARAGRAPH
When lit, U4 and U5 produce on the order of 100uV of
offset voltage, with source negative and drain positive.
Asymmetrical resistors R14 and R15 forming the common
point between U4 and U5 transmit the respective offsets
unequally, resulting in a net positive, which current
from U3 via R6 and R7 trims back to zero. The trim
current is adjustable from 1.6uA to 4.4uA, developing
16 to 44uV across R114.
NEW PARAGRAPH - February 2020 - ALSO DISCREDITED - January 2021
U4 and U5 are selected so their mostly cancelling offsets
result in a net positive, which current
Oh, and one more mistake I made. R3 is marked 10031, not 20011. Must be 10K, since 130 ohms is absurd. That means current into U3 is (15-2.5) or 1.25mA .
I don't know what I was thinking.
I sure wish I knew U3's part number. I want to say LM336, but 162K for R5 sounds absurd.
Thanks,
Dave Wise
UPDATE 2021-Jan-25
When I built the H11F1 modulator for my 740B #1, I matched offset. This turned out to be a mistake, and I rebuilt it with a new round of selection and matching.
Using a toaster oven environmental chamber, I recorded the offset voltage of nine H11F2, all the same production lot.
I recorded V at 15mA, my chosen drive current, and 20C, 30C, 40C, 50C, and 60C, then calculated TC. I also varied the drive current and calculated drive current coefficient IC.
Offset was all over the place, some positive some negative, but all had positive (and fairly constant) TC and IC. I found two good matches.
TC was approximately +0.5uV/C and IC +0.5uV/mA. (Pin 6 with respect to pin 4.) In my chosen pair, one had positive offset, one negative. The sum when wired in series pin 4 to pin 4, i.e. cancelling TC and IC, was around 50uV.
This is more than the zero adjust range, so like Fluke I added a coarse trim. In the 740B this is easy, just insert a resistor at the appropriate end of the control.
In the oven, my chosen pair drifts only a couple uV from 20C to 60C.
Note from TiN's picture in post #21 that U5 has a paint dot. I believe this reflects selection and matching, and dictates install order so the net offset is positive. Which is then cancelled by Fluke's negative-only coarse trim.
NEW PARAGRAPH - January 2021
When lit, H11F optoisolators produce significant offset voltage.
U4 and U5 are selected so their temperature coefficients cancel.
Secondary selection rejects pairs with excessive total offset.
Then the pair is installed such that the offset - U4 pin 6 minus
U5 pin 6 - is positive. Current from U3 via R6 and R7 trims this
back to zero. The trim current is adjustable from 1.6uA to 4.4uA,
developing 16uV to 44uV across R114.