Andreas, DiligentMinds, branadic
up to now, this thread was a very instructional and successful demonstration, how such delicate measurements like the determination of these low T.C. of resistors could be done with relative simple means, instead of using heavy equipment like an 8.5 digit multimeter, or ultra stable and precise resistance bridges..
But at that point, it is no way of judging, which resistor technology or brand is better, or best in class..
Reason for that is the mere lack of statistics.. means, too few different samples per technology / brand have been used.
Elsewhere, I already have provided 5 T.C. values for the hermetically sealed, oil filled VHP202Z, and found out, that these values showed a big spread, and all were an order of magnitude above the promoted typical value.
(Andreas, if you like, I may add these measurements here as an additional item)
To evaluate, what the different technologies are really capable of, say in statistical terms of 'typical' values and max. spread, you need much more samples than 2, or even 5 ones.
Therefore, up to now I would not draw such a conclusion, as DiligentMinds did.
Maybe others also can add such measurements...
Andreas, if you don't have an 8E16 resistor yet, we may discuss, how to organize that.. I have a bunch of 50 Ohm, .., 10k, 30k, 100k in stock.
Either I do some much better measurements on them, (I mean, than the quick *n dirty one that I already did
) using my equipment, or we may exchange some samples..
The basic problem with these tubular case resistors is that the wires obviously transport most of the heat into the resistor, compared to the mould component.
That effect can be reduced by attaching the leads thermally to the isothermal block also.. there are several possibilities to do so.
Anyhow, this subject is still very interesting, so keep on going!
Frank