Hi Alan,
No, issue I have is not temperature drift. I spent days to make sure of it. And I'm very confident it's not drift of A9 LTZ reference output or something on A1 DC PCBA either.
I did left it drifting for few days before, and it had constant rate, going over 20ppm off, which is over any possible spec. K2001 connected to same source was showing stability in ppm level.
Anyhow,
ManateeMafia had multiple 3458A's sampling 10V, and I finally got ahold to tidy-up and make page with summary graphs -
4 x HP3458A test of Fluke 732B. It's permalink
https://xdevs.com/datalog_comp/, updated in realtime every 5 minutes, as new data comes in.
There are two caveats before you can study data graphs:
1. Sometimes data CSV's had delays, but graph shows as interpolated line between. For example there is no data at December 31, but line is there.
2. Time/date scale axis is not same on all graphs (as different meters were added into measurement at different days)
I don't know JavaScript or D3.js to figure how to fix these issues at this moment, so you warned
Lessons I've learn't so far from data:
Meter 04188 have drift
+0.066ppm/day and tempco
+0.066ppm/KMeter 08451 (with real 002 option) have drift
+0.145ppm/day and tempco
-0.266ppm/KMeter 01527 have drift or tempco
4ppm/day?. Not enough data to be sure (only 2 days history so far)
Meter 02157 with repaired A3 ADC have drift
-0.05ppm/day and tempco
-0.266ppm/KMeter 02157 with good A3 ADC have very low drift (not much data to say for sure). Note finer ppm scale
I think this is very good data to answer question how good stability 3458A should look like...
Humidity, pressure and ambient temperature are also logged by standalone sensor hooked to same Pi, so you can nit-pick if there is any relation...