Author Topic: Repair worklog : Old rusty HP 3458A  (Read 252928 times)
Offline TiNTopic starter
  • Country: ua
    • xDevs.com
Re: Repair : Old rusty HP 3458A
« Reply #350 on: November 17, 2015, 01:01:32 am »
VK5RC
If you asking HPAK to repair your 3458A, you doing it wrong  :-DD.
Welcome to high-end gear. It's "normal" out there.
Macbeth
Yes, but no. We want perfect, right? How much perfect is only matter of disposable funds available...
Drift ain't there now  :rant: overnight span ~2.4ppm (temp change 26.1C to 30.3C), so that's OK.
I have wild theory now why it's not drifting now. No, A3 not changed, I did not even unscrew bottom cover yesterday. You guess? :D
Offline ManateeMafia
« Reply #351 on: November 17, 2015, 01:10:16 am »
Air filter?
Offline Macbeth
« Reply #352 on: November 17, 2015, 02:12:07 am »
I know, I know!!! That bloody cat is opening doors and letting the drafts in!  :-DD
  • Country: ua
    • xDevs.com
« Reply #353 on: November 17, 2015, 04:08:02 am »
No, not so obvious as air filter.
And no pets allowed in volt-nut house  :-DD.
Hint - it's not temperature or voltage related.  ;)
Offline Vgkid
« Reply #354 on: November 17, 2015, 04:34:49 am »
Bad power connection/cable.
  • Country: ua
    • xDevs.com
« Reply #355 on: November 17, 2015, 04:40:37 am »
Nop. Nothing obvious, this one is tricky.
Offline crispy_tofu
« Reply #356 on: November 17, 2015, 06:11:55 am »
I was thinking something along the lines of a wobbly connection in the banana inputs, but it's not obvious...
Dodgy software?  :-//
Offline VK5RC
« Reply #357 on: November 17, 2015, 07:32:00 am »
?Dissimilar metals of connection cable/connectors??
RE HP-A-KS just thought I would phone them to see how much in AUS, it is part of my tendency to "nuttery" voltage or time!!!! :scared:
Reminds me of the joke "a man with one clock knows the time, a man with two is not so sure" perhaps "a man with one DMM knows the voltage……. etc"
I have a slightly bulgy main 15000uF in the main PSU, a replacement is coming soon.
Online deadlylover
« Reply #358 on: November 17, 2015, 07:55:19 am »
Hmm I might replace all electrolytics, because once you get it calibrated you don't want to ever open the thing up again until it requires repair.
Humidity/moisture in parts being released after a long cycle?  ^-^
Offline plesa
« Reply #359 on: November 17, 2015, 08:49:54 am »
My guess - thermal EMF
  • Country: ua
    • xDevs.com
« Reply #360 on: November 17, 2015, 10:05:06 am »
plesa is close, but no, not EMF :) Not related to thermals.  >:D
Offline Monittosan
« Reply #361 on: November 17, 2015, 10:13:22 am »
If its ok by TiN pics pics pics! I had to stop my self from opening your 3458A soooo many times just to keep the cal seal intact and let the next owner have the satisfaction of cracking it open  ;D
Offline Dr. Frank
« Reply #362 on: November 17, 2015, 10:53:30 am »
Maybe, for this missing shorting jumper on jack P100, you used a TiNned one, instead of gold plated?
Oxydation might create unwanted chemical galvanic element in the signal path.
You know, P100 near the A9 PCB, which selects GND, +7V, or Normal, and which you might have used (and touched) for measuring A9 output.
Frank
« Last Edit: November 17, 2015, 12:03:47 pm by Dr. Frank »
« Reply #363 on: November 17, 2015, 01:36:26 pm »
Dr Frank,
 That would be a good catch even if it wasn't the main problem. At least TiN has a donor board to get a jumper.
Offline Dr. Frank
« Reply #364 on: November 17, 2015, 02:13:04 pm »
Dr Frank,
That jumper already made me nervous when I have seen it for the first time.
On 2nd thought, it may be inside the AZERO circuit, and so any parasitic voltage would be eliminated.
Frank
Offline plesa
« Reply #365 on: November 17, 2015, 03:00:20 pm »
I must give up :-//
Offline WN1X
« Reply #366 on: November 17, 2015, 03:31:40 pm »
I am going to guess it is light leakage affecting some of the IC's.
- Jim
Online nfmax
« Reply #367 on: November 17, 2015, 03:37:13 pm »
If all else fails, blame the software ;)
« Reply #368 on: November 17, 2015, 03:46:22 pm »
Test gear or other device with fan blowing warm air into rear of meter.
Online Zucca
  • EE meid in Itali
« Reply #369 on: November 17, 2015, 03:52:52 pm »
I must give up :-//
You are not alone, my head is still smoking and nothing is coming out.
PS: I imagine TiN with a cold beer smiling on us  ;)
« Reply #370 on: November 17, 2015, 04:02:51 pm »
I would be smiling too if I knew that A3 board was working!  :phew:
Offline Macbeth
« Reply #371 on: November 17, 2015, 04:46:59 pm »
That's it! It's the bloody compressor cycling on his beer cooler! Should have got a peltier cooler instead.
Offline teksturi
« Reply #372 on: November 17, 2015, 04:48:45 pm »
Power line frequency is not correct or stable or something similar.
Offline WN1X
« Reply #373 on: November 17, 2015, 07:08:23 pm »
The compressor is cycling too much because there is no thermal load for it after TiN drank all of the beer  :-DD
- Jim
Offline Macbeth
« Reply #374 on: November 17, 2015, 07:41:33 pm »
I noticed a few bottles of India Pale Ale during his clean up operation. That Taiwanese IPA is nothing like the stuff over here, I mean 5-6% abv is normal, but 99%?  :-DD
