Meanwhile the meter board is partially populated and i made a power supply with a shielded mains transformer: -33 V, -12 V, -5 V, +3,3 V, +5 V, +12 V, +18 V). The design is such as to run everything with minimum heat. 3.3 V is for the FPGA module that includes switchers for 1.2 V and the 2.5 V needed for the LTC2386 digital section. We'll see whether the switchers are a problem.
The dual LTZ1000 reference works as intended with VRef = 14.1876 V and about 5V heater voltage at ambient temperature. The negative reference (same as Prema 6048) is -1.75 V. The PI temperature controller got configured for 28 °C. The front end with the BS supply works, too. It went stable after adding the 470R + 1nF filter shown in imos recent simulations. When feeding the meter input with low impedance, output rise/fall times are about 50 nsec with little overshoot. With the standard input protection and filter this increases to about 1.2 usec.
Regards, Dieter