Hello Friends,
I purchased a Keithley 155, and now I’m trying to get it working correctly. When first received, the needle pinned to one side on any range. I started by replacing the electrolytics, and many of the carbon resistors. A lot of the carbon resistors had drifted 30-60% out of range. After doing those repairs, the DC voltages are on point. The multivibrator circuit is working perfectly, and putting out the square wave specified in the manual. It has been adjusted to the proper 220 Hz. Everything is looking good, but now there is no meter movement with voltage on the input.
The meter itself definitely works, as the unit has battery check function, and the needle pins to either side for the positive and negative voltage check. I started going through the circuit, and the problem seems to be somewhere in the amplifier section (I think). I’ll show some of the waveforms & measurements further down, what I’m looking for some help with is parts replacement suggestions for some of the transistors that don’t show a replacement in the manual, they are listed simply as “Selected at Factory”. Also any tips on debugging the problem are more than welcome.
Here’s the schematic:
Parts List:
Q101 & Q102 on their body have text on them: “HDIG1030”, which apparently can be replaced by 3N163. The JFET Q118 is listed as 2N5457, which I was able to find some of online.
Do you guys have any replacement suggestions for these parts: Q103, Q106, Q107, Q108?
Here's some measurements I've taken below-
Multivibrator waveform at the gates of Q101 & Q102:
Output of Q102 to C105:
R148 into C112:
The other side of C112 seems to be mostly shorted by Q118, unless I'm reading this wrong
Have any ideas?
Also see attached file of K155 board that is has much more clear reference designators than the manual