Author Topic: optimal configuration of HP 3458A to measure 10k  (Read 10620 times)
Offline e61_philTopic starter
« on: March 01, 2018, 09:16:19 am »
I know, we discussed that already severeal times, but it is hidden in various threads.
So the question is: "How to configure the 3458A to get the most accurate 10k reading"
My current configuration is the following:
OHMF
RANGE 10000
OCOMP ON
DELAY 2
APER 1
NDIG 8
Is there anything to improve?
I take several measurements with this configuration and calculate the average.
Best
Philipp
Offline TiN
  • Country: ua
    • xDevs.com
Re: optimal configuration of HP 3458A to measure 10k
« Reply #1 on: March 01, 2018, 10:24:39 am »
DELAY depends on cable, some cables might need as high as 5, but most high quality cables are fine with 1-3 sec.
you may want to NULL meter with short prior to DUT if possible. Or not :)
Also beware tempco, 3458A (at least mine) are not that great in this respect, so you either run ACAL every 0.5c or characterize tempco of the meter itself by using thermally anchored DUT resistor first.
Keithley 2002 is much better for tempco in this respect on 10K range.
« Last Edit: March 01, 2018, 12:30:42 pm by TiN »
Offline Dr. Frank
« Reply #2 on: March 01, 2018, 11:20:19 am »
Singular measurement samples will have too big an error, as the noise / standard deviation in Ohm mode is at least 0.2ppm of range.
Therefore it's mandatory to invoke the statistics function for measurements.
By measuring the StD also, you will have control over possible erroneous measurements, as can happen easily by external disturbance. If you observe an excessive StD value, just skip the measurement.
With these above settings, take 16 readings and use mean value for the Ohm value, and the StD value as its uncertainty value (plus calibration uncertainty, of course).
The setting for my 10k resistors and cabling, and the 'workflow' for taking measurements is therefore:
OHMF
RANGE 10000
NDIG 8
OCOMP ON
DELAY 1
APER 1
NRDGS 16
TRIG HOLD
MATH STAT
TRIG SGL or TRIG
RMATH NSAMP
RMATH MEAN
RMATH SDEV
I have stored the settings in the first block to a State, to be accessed by e.g. RSTATE 2, and the MATH functions are assigned to several Keys, copied that configuration from the FLUKE  3458A/HFL.
A single measurement then lasts about one minute.
Once configured that way, you can easily repeat the measurement, or measure the next resistor by pressing
MATH STAT
TRIG
PS:
MATH NULL is not recommended, because OCOMP and 4W function already compensate NULL errors. Additionally, for using NULL, you would have to change the cabling configuration and would probably subtract a residual Offset which has nothing to do with your 10k DUT.
Also, MATH NULL will delete the STAT function, and it's difficult to invoke both functions in a logically correct manner.
Frank
« Last Edit: March 01, 2018, 11:43:37 am by Dr. Frank »
The following users thanked this post: quarks, TiN, franklin
Offline PartialDischarge
« Reply #3 on: March 01, 2018, 11:27:41 am »
Add AZERO ON
Offline Dr. Frank
« Reply #4 on: March 01, 2018, 11:32:22 am »
Add AZERO ON
AZERO is on by default, but is not necessary due to OCOMP ON, as this automatically subtracts  voltage offsets from the DUTs wires, but also internal voltage offsets.
If you select OCOMP ON, the 3458A does not use a dedicated AZERO cycle any more, I strongly suppose, because the measurement time is not increased over about 4 seconds / sample.
« Last Edit: March 01, 2018, 11:49:24 am by Dr. Frank »
« Reply #5 on: March 01, 2018, 11:44:24 am »
True
but is not necessary due to OCOMP ON, as this automatically subtracts  Offsets, from DUTs wires, but also internal voltage offsets.
If you select OCOMP ON, AZERO is not used by the 3458A any more, I suppose, because the measurement time is not increased over about 4 seconds / sample.
Are you sure?? I'm not that sure about that, and the manual does not suggest that exclusive behaviour anywhere. Someone even posted this question regarding his suspicion on the inner workings of OCOMP and AZERO
https://community.keysight.com/thread/20408
« Last Edit: March 01, 2018, 11:46:43 am by MasterTech »
Offline Dr. Frank
« Reply #6 on: March 01, 2018, 11:53:32 am »
True
I'm pretty sure, because I wondered, why the sampling time did not double by using OCOMP ON.
The manual is lacking some information at several instances
« Reply #7 on: March 01, 2018, 12:10:11 pm »
Thanks for your comments!
It seems to be already the right setup, that I used.
I do all statistics on the computer.
As an example I attached a measurement of my SR1010-1k Transfer Standard (R1-R10).
Std. Deviation is 0.38ppm over 1000 Measurements
Temperature was well within 22°C +/- 0.1°C
Best
Philipp
« Last Edit: March 01, 2018, 12:18:25 pm by e61_phil »
Offline Alex Nikitin
  • Femtoampnut and Tapehead.
    • A.N.T. Audio
« Reply #8 on: March 01, 2018, 12:43:09 pm »
Yes, the tempco of our 3458A on 10K is somewhere between 1 and 2 ppm/C  :( .  It is about the same on my 3456A meter at 10K.
Cheers
Alex
« Last Edit: March 01, 2018, 12:44:58 pm by Alex Nikitin »
Offline Dr. Frank
« Reply #9 on: March 01, 2018, 01:08:01 pm »
Cheers
Alex
I assume, that's the T.C. w/o ACAL?
It's specified 3.1ppm/K,  but 1ppm/K with (or right after) ACAL
Frank
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #10 on: March 01, 2018, 01:15:50 pm »
Yes, of course. ACAL helps, but not very practical if ambient is unstable due to 860 seconds wait time.
« Reply #11 on: March 01, 2018, 01:26:42 pm »
Why is the K2002 that much better? Is the VHP101-40k only used for the ACAL and not for 10k in the 3458A? Or is it the circuit around?
And how is it done in the K2002?
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #12 on: March 01, 2018, 01:43:32 pm »
It's kinda obvious if one look at 2002 current source and 3458A current source schematics and resistors used.
And no, 40K not used in resistance measurement directly.
The following users thanked this post: e61_phil
    • A.N.T. Audio
« Reply #13 on: March 01, 2018, 01:53:05 pm »
Cheers
Alex
Frank
Yes, without ACAL. ACAL helps but doesn't remove the temperature dependency completely (so even after ACAL the measured value somewhat depends on the ambient temperature).
Cheers
Alex
Offline dl1640
  • Country: cn
« Reply #14 on: March 13, 2018, 12:35:59 pm »
I'd rather wonder how to make a correct null on 3458 before measuring a voltage source, 1) null by copper short on 3458 binding posts, 2) null by shorting measurement leads, 3) null by measuring voltage source set at 0V and output on, 4) dont null..
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #15 on: March 13, 2018, 01:20:52 pm »
Depends on what you trying to do. Usually it's 3.
Offline CalMachine
  • Metrology Nut
« Reply #16 on: March 13, 2018, 01:40:24 pm »
Depends on where you want your plane of reference to be.  At the end of your leads?  At the front of your meter?  At the front of your source?
All your volts are belong to me
Offline cellularmitosis
« Reply #17 on: March 13, 2018, 11:09:30 pm »
What might determine that?  Perhaps, "Am I shipping the cables with this reference, or not?"
LTZs: KX FX MX CX PX Frank A9 QX
Offline CalMachine
  • Metrology Nut
« Reply #18 on: March 14, 2018, 12:25:04 am »
Generally personal preference.
Offline dl1640
  • Country: cn
« Reply #19 on: March 14, 2018, 12:35:17 am »
IMO it is good to use method 3 if check "gain", but to check "zero" output from the source, one would consider to null out any cabling effect and/or DMM "residual" reading beforehand. Some would use method 2, to short two leads and let them hanging in the air, or short them at either of the source terminal, then take a null, then change the cable config to gain measurement, this means the cable config at null measurement and gain measurement are diffierent, I am not sure if such null is correct for precise application. Most of time I do not have to measure down to a few or sub ppm, so I do not null, method 4.
Please correct me if I'm wrong.
Also, I think method 1 can be used to check the "true" null of DMM, but need a good copper short.
Offline CalMachine
  • Metrology Nut
« Reply #20 on: March 14, 2018, 02:11:39 am »
Yes, as you pointed out, each scenario is applicable to certain situations!
Offline dl1640
  • Country: cn
« Reply #21 on: March 14, 2018, 03:15:29 am »
For precise DCV measurement, I would consider a remote controlled switch to form and release a short at DUT end. This is method 2.
For Ohm measurement, I use 4-wire, OCOMP ON, and Delay 1, NPLC 20~100.
« Reply #22 on: March 14, 2018, 05:12:14 pm »
This controlled switch should be very good. Most relay contacts I know are specified with around 2µV.
Offline dl1640
  • Country: cn
« Reply #23 on: March 15, 2018, 12:22:20 am »
Or, if the source has binding posts, then just leave the cable config as gain measurement (Hi-Hi, Lo-Lo), and place a good copper short in between the binding posts, I think this is a good null, then remove copper short and measure gain.
Offline voltampere
« Reply #24 on: April 11, 2018, 06:50:50 pm »
I hope the answer to my question is not just a trivial thing that i could have found in the manual myself!
I am scratching my head regarding the use of the APER vs the NPLC command.
According the manual it should have the same rersult! ( as far as i understood it ).
Meaning if i set APER 1 then the integration time should be 1 second.
And if i say NPLC 50 then it should also be 1 sec, right ( on a 50 Hz model ).
I tried the APER command like this
DELAY 1
APER 1
NRDGS 15
And it will take roughly 89 sec to complete the 15 readings. ( hand measured, +/- 1 sec ).
If i use
Delay 1
NPLC 50
NRDGS 15
then it will take roughly 204 sec to complete 15 readings.
OCOMP ON for both sessions.
It is the first time i used the APER command because Dr. Frank mentioned it in his reply!
Furthermore:
If i command APER 1
then i get the following results when querying:
NPLC ? -> 50
APER ? -> 1.0 E+0
If i command NPLC 50
i get the following results:
NPLC ? -> 50.0 E+0
APER ? -> 200.0 E-03
I hope somebody can shine a light for me!
Regards
Joerg
EDIT:
Typo ; it looked like i used the same commands; instead it should read NPLC 50 on the second run ( with 204 sec )
« Last Edit: April 22, 2018, 08:32:57 am by voltampere »