Author Topic: Old fashioned zener 10V reference  (Read 41187 times)
Offline Conrad HoffmanTopic starter
    • The Messy Basement
Re: Old fashioned zener 10V reference
« Reply #25 on: February 17, 2022, 12:24:53 am »
Just guessing, but I have to believe an opamp current source would beat any simple discrete circuit. Still needs a reference in there somewhere!
Offline Tj138waterboy
« Reply #26 on: February 17, 2022, 12:33:35 am »
I was thinking the same but all schematics of older voltage calibrators such as edc mv106, rv622 use bjt current mirrors when opamps were definitely available.
Offline Kleinstein
« Reply #27 on: February 17, 2022, 07:43:16 am »
The circuit like in the original post is a very stable current source. The current is set by the resistors (one ratio for the gain setting resistors and the one from the 10 V to the zener) and the voltage reference. A normal transistor based current mirror is not really that accurate.
In the old time it was common to use 2 zeners in a kind of bridge circuit, so that one zener would set the current of the other and the sum of the voltages would be the final output. In case one wants a 13 V reference, this may still be an option.
For the temperature stability one could with most diodes trim the linear TC via the zener current, but the 2nd order TC still remains. If really needed one may combine this with a regulated temperature. With a relatively low TC to start with the temperature regulation does not have to be perfect. I would consider a regulated temperature simpler than the 2nd order TC compensation circuit used in the Solartron 7071/7081.
The following users thanked this post: Tj138waterboy
Offline guenthert
« Reply #28 on: February 17, 2022, 05:01:35 pm »
Have to resurrect old post to ask if one wanted to make a good current source to tempco test a 1n829, which circuit  construction would be more temperature stable? Opamp current source, wildar current source 2 or 3 transistor version or wilson current mirror 3 or 4 transistor version, or would basic single transistor current source be just as good? Will be powered by stable dc power supply hp6111a @ 12v. Will be cycling temperature of only 1n829a zener attached by wirewraped leads inside insulated cookie tin.  Current control circuit will be built on solderless breadboard or project board in room that temperature varies between 50-65degrees.
     50 years ago, H.A. Cole at Cern designed a voltage reference for wide temperature ranges.  He used William's ring-of-two reference as constant current source to drive a temperature stabilized Zener (1N825 -- I believe that's the one used in HP735A).
http://cds.cern.ch/record/1063330?ln=en
« Last Edit: February 17, 2022, 05:07:39 pm by guenthert »
The following users thanked this post: 2N3055, Tj138waterboy
Offline dietert1
  • Country: br
    • CADT Homepage
« Reply #29 on: February 17, 2022, 05:33:25 pm »
In 2009 i made a 10 V reference with a AD587 (small Geller board). That IC included a buried zener, a precision resistor network and an opamp. One could connect a capacitor as noise filter. I glued an NTC on top and made an analog PI temperature controller with a TL071 and several hundred uF for the integrator. Few parts, nothing special except the reference itself.
The reference was good enough to check 6.5 voltmeters. Observed drift was some ppm over 10 years, in comparison to a pool of well aged HP 3456As.
Regards, Dieter
Offline Gyro
« Reply #30 on: February 17, 2022, 08:19:50 pm »
It's a real shame that Joe Geller wasn't able to justify continuing to produce those boards and keeping his Fluke reference in Cal. I was lucky enough to purchase an SVR-T - the one with the thermistor based second order compensation that Lars came up with. He had a really neat micro thermal chamber for sweeping and compensating them. It spends most of its time turned off, so is hopefully still within a few ppm.
I still kick myself for not buying some extra 586 and 587LQs, and other stuff, when Joe was selling off his stock!
Best Regards, Chris
« Reply #31 on: February 17, 2022, 11:31:33 pm »
That is a great read. Especially the findings that varying voltage 20-24v and 24-40 resulted in positive and negative voltage tempco. I believe that I may have came across this circuit in a Keithley schematic before but didn't understand what was happening with the use of 2 other zeners within the current mirror. Just assumed I was to provide a stable voltage for the current but now realize it more than that.
« Last Edit: February 17, 2022, 11:36:43 pm by Tj138waterboy »
Offline TimFox
« Reply #32 on: February 18, 2022, 12:31:58 am »
A trick used in some older precision power supplies, that had a TC Zener in a small oven, was to connect the Zener and three resistors in a bridge circuit, that could be adjusted to cancel out the Zener's resistance by adjusting the bridge for maximum voltage.  I thought I had an example in an old Motorola diode handbook, but I can't locate it.
« Reply #33 on: January 23, 2023, 10:46:29 pm »
Finally getting back to test setup after losing my jantx1n829a I found a lot of 10pc 1n825 and have been looking at different setups suggested and wanted to know if there is anyone that can see a problem with the way im looking at setting up a fixture to burn-in all 10 refs at the same time. Given The limitations of 5.5 digit hp3478a, mv106j dc voltage standard calibrator, and fluke 840 nA galvanometer it would seem to make more sense to null the references to calibrator maybe with a mux. The current configuration is feeding 15v into a lm7812 getting output of 11.8v. All 3904 transistors were matched to within 5mV vbe. The goal is to get the recommended 7.5mA to all zeners simplest and most accurate way possible which is why I am currently using a basic multi-stage current mirror1698526-0 I dont know if this has any drawbacks for my specific use purpose but am using a free circuit simulator "Every Circuit" to try to make the Improved Wilson current mirror into multistage and wanted to see if someone thats more knowledgable can tell me if my implementation is correct and or if theres a simpler/better way to make 10 mostly precision tracking current source from opamp better or cheaper. All resistors are 50ppm or better values are exactly as my test setup.
« Last Edit: January 23, 2023, 10:49:41 pm by Tj138waterboy »
Offline Andreas
« Reply #34 on: January 24, 2023, 08:15:43 am »
Hello,
for a simple burn in cirquit I cannot see the advantage of having exactly 7.5 mA.
If I remember right from the discussions on Volt-Nuts the burn in was done at much higher currents (more like 40-50 mA)
(Google for "volt nuts some questions to zeners". I think it was user WarrenS which had a lot of experience)
After burn in the zener was not operated at 7.5 mA but on the "Zero TC" current which is individual for each zener (at a specified environment temperature) and usually in the range of a factor 2 around the 7.5 mA.
So the 7.5 mA is only a test current to select wether the zener is a 1N821 or a 1N829 at 25 deg C.
Essential is also to screen the zeners for "popcorn noise" to sort out the "stinkers".
After my experience the 1N82x sold nowadays have relative high 1/f noise (in relation to the high current).
So I stopped my experiments with 1N829.
with best regards
Andreas
« Reply #35 on: January 24, 2023, 12:28:52 pm »
Im sorry for not detailing but the circuit is adjustable current mirror so i can adjust from around 6.5mA -  20mA. With the mirrors all tracking each other it should be quicker to find the better zeners per specific current vs testing individually. I will have a look into burning in at higher current thanks for the info.
Offline guenthert
« Reply #36 on: January 24, 2023, 06:19:55 pm »
   As others have already written, the Wilson current mirror isn't expected to deliver exactly the same current nor is the 7.5mA necessarily the right current.  I just can't help, but stating the obvious, that to get the exact same current to all Zeners they ought to be in series (string of Zeners).  That might actually be a reasonable approach to a 100V reverence as it reduces noise by n1/2.
« Reply #37 on: January 25, 2023, 12:48:45 am »
Thanks Andreas, I did come across that exact article you mentioned previously which is why I decided to go with a batch of 1n825s instead on 1 or 2 1n829s. My current set resistor can easily be modified to lower the current to start from 4-5mA. I am just a hobbyist without any schooling on e.e. subjects and just want to build, take apart and learn as I go. My post was more asking if there are any issues noticeable with the construction and simulation of designed multistage improved current mirror due to I have never designed my own circuit only copied others and modified values. Its obvious that all mainstream variations of current mirrors work but there is no data I can find that anyone has built or tested the version I have simulated. Even if using a cascode setup adapted to the design could be useful or if I better match transistors vbe would be worthwhile. I understand this is in the metrology section way above where this probably should be asked but we all have to start somewhere. Thanks for all the knowledge you guys share.
Offline Vtile
  • Ingineer
« Reply #38 on: February 04, 2024, 12:56:22 am »
This was super fun build, I did one years ago. Today I powered it up again after it have been laying on under a pile of trash above the workbench for several years.. After an hour warm up, both my Keithley 197A (LM399) and the 1N823+OP07 reference the value set to 9.9999V which is not bad, it still is "too good" for the best measuring equipment I have the 197A.
The current source I was using were simple.. Forementioned Keithley as current meter in series with some 10 turn potentiometer, powered with Thandar lab supply (or were I driving the PSU against current limiter and using it as current source... can't recal). Then my MeraTester v640 (100meg, 1.5mV FSD) as a null detector against another channel of close enough voltage from Ch2 from thandar supply. Then the heat was cycled first by pinching the dut (1n823), then adjusting the current in a point where no movement of null detector was seen.. After that I did use a 12V lightbulb wrapped to foiltape (to allow only heat) put under and contact to the DUT,  then again adjusting the current to null out the voltage on null detector on this rather violent heatcycle.. This also involved to disassemple the current measuring setup and to use the Keithley 197A to measure the TC voltage after point was found. Repeated this to three 1N823s and did choose the best one (can't recall did I choose the one with widest TC region or something can't recal)..
Then calculated the needed resistance between the 1N823 and 10V output.
Calculated the needed resistance for amplification multiplier resistors.
Sorted out all my handfull of 15ppm 0.1% resistors for TC. This was done again with Thandar as supply and 4 of same value resistors as Weatstone bridge configuration. One as a dut and 3 others as voltage source for null meter. Dut temperature cycled with my fingers and variation detected with zeroed MeraTester v640 in lovest range possible (mostly 1.5mV). I might have used a 10 turn Bourns pot again at one leg of the bridge.. Can't remember all details, it was a mock up one weekend setup, with a few beers.
Calculated series parallel networks to match the needed resistor values and then
constructed the circuit in manhattan style to full copperclad board with attention for mechanical stresses.
As an afterthough I should have put more efford to the amplification multiplier resistor network it is sub optimal part.
« Last Edit: February 04, 2024, 01:18:55 am by Vtile »
The following users thanked this post: Andreas, Victorman222
« Reply #39 on: February 04, 2024, 03:25:47 pm »
Wow, an oldie, but I still have that board. I think an opamp current source would be best but I'm wondering if you can use the DUT for the current source and measure the output, sort of combining the whole thing. No specifics, just armchair pondering. For TC on small parts I use a couple small beakers of mineral oil and just dunk them. Very fast, though you might want a small lab hotplate or immersion heater (or a medium sized low value resistor.) to keep things warm.
Offline Victorman222
  • Country: ua
« Reply #40 on: February 04, 2024, 05:19:19 pm »
Yes, Conrad suggested the same thing to me along with measuring change relative to another reference and it worked well, i was able to characterize some soviet 9V reference diodes to <2 ppm/K over 10 degrees with dt830 multimeter.
I added multiturn potentiometer to control the current of the diode and the output voltage, the diode was on long wires fixed to a light bulb with a rubber band, measurement was done relative to same circuit but with fixed resistors. The graphs in the Microsemi datasheets helped me get my bearings showing the relation of TC to current and zener voltage to current.
Thanks everyone for the great info in this topic!
Offline Kleinstein
« Reply #41 on: February 04, 2024, 05:36:09 pm »
The trimmers are OK to find the right gain and right amplification. However even good trimmers are usually not very stable over long time. So one should get at least most the resistance from a set of fixed resistors and a trimmer (usually a higher resistance like ~10 K) only for a small part. The is especially for the gain.
The following users thanked this post: Victorman222
  • Country: ua
« Reply #42 on: February 04, 2024, 06:01:19 pm »
Yep, the trimmer circuit is only to find the 0 TC current. Those soviet WW trimmers are something like 500 ppm/C so it's far from a reference. I wrote down the resistance values into an excel sheet and when it came time to using the diode i put the appropriate current setting resistor.
Offline VE2UM
« Reply #43 on: February 04, 2024, 08:31:32 pm »
excellent idea about the 1N829.
However, you’ll have to replace the two 15k resistors with a 620 ohm device.
1N82x zeners are rated at 7.5 mA bias..
Also, may be a little pullup from the Supply to the opamp output to aid it supply this 7.5 mA 1N829 bias.
Offline EC8010
« Reply #44 on: February 05, 2024, 01:16:09 pm »
I wasted ten minutes trying to find NS AN161 on the Internet. Then I did what I should have done first and looked in my library, finding it in the National Semiconductor Linear Applications Handbook. Doh!
Although an op-amp CCS is simple and usually best, sometimes you're forced to use discrete component circuits such as the Williams ring-of-two.
Another way to achieve temperature compensation is a string of references. Attachment shows the effect of connecting a string of BZX55C5V6, some in forward, some in reverse. The diodes were immersed in white oil.
Offline alm
« Reply #45 on: February 05, 2024, 01:44:34 pm »
Here is an electronic copy for anyone else reading along: National AN-161: IC Voltage Reference has 1 ppm per Degree Drift.
The following users thanked this post: EC8010
Offline Vtile
  • Ingineer
« Reply #46 on: February 05, 2024, 10:56:25 pm »
This is actually a really good project, not just for hobbyist metrology wannabes (like myself), but to get a stable and relatively (take your best voltage measurement device and call it done) known voltage source for the hobby lab. Output of the reference could be fitted with eg. 25 turn potentiometer which output is then puffered with opamp set as voltage follower (buffer) and/or voltage to current converter. As a single precision potentiometer used as (... eh, well..) potentiometer the tempco shouldn't be that much of an issue since the material is homogenous both sides of divisor. Buffer then makes this arrangement as "ideal potentiometer" as they were called in 70s.
I had originally idea of building three pieces of these and my intention was to make this kind of arrangement to get more stability and controllability for voltage and current for characterise the two or three other zeners and other parts. Only issue is that the first one is too good to begin with and I would need to upgrade to 6.5 digit meter to keep interest. Haha.
« Last Edit: February 05, 2024, 11:16:55 pm by Vtile »
Offline Rax
« Reply #47 on: February 06, 2024, 04:24:53 am »
"My "oven" was nothing more than a cardboard box over the board with a lightbulb shining on it. Warm the board up a dozen degrees, cool it down (all slowly) while monitoring the voltage.
Actually, great idea. Simple, easy to implement. May steal it?....
Needing to find tempco resistor for an LTZ1000 kit.
Offline Andreas
« Reply #48 on: February 06, 2024, 07:46:41 am »
Hello,
why not measure the difference of 2 references in 100 mV range and optimize further.
(simply use your DMM as null-Voltmeter).
Most 6.5 digit multimeters deliver more resolution than that displayed in default mode.
Just use the math function or the interface to get higher resolution.
So you can still have a lot of fun before you need a higher resolution meter.
https://www.eevblog.com/forum/metrology/getting-one-more-digit-from-a-6-5-digit-meter-without-using-gpib/msg1300529/#msg1300529
with best regards
Andreas
Offline iMo
  • Country: bt
« Reply #49 on: February 06, 2024, 10:07:54 am »
Hello,
(simply use your DMM as null-Voltmeter)..
That IS the best method actually :)
My first measurements were made with a diy nullmeter (LTC1050 opamp with a 100-1000x gain) wired floating between two references, with a 10turn ww pot to get them nulled somehow. The dmm was a 3.5digit meter, and I saw X uVolts drifting as well  8)
« Last Edit: February 06, 2024, 10:10:58 am by iMo »