The reason I said "without adding more units" because I have seen several teardowns for high-end benchtop multimeters, 6.5 to be specific that have only one LM399 reference, so I thought there is no need to have extra LM399 and maybe there are alternative ways to just reduce the noise with maybe just active filtering or something similar!
Hello,
its not active filtering but selection.
filtering is difficult at frequencies near / below 0.1 Hz.
You would need large (electrolytic) capacitors which have
(not constant over time) leakage currents that also create noise.
And you do not want a "stinker" LM399 with > 7uVpp in a precision instrument.
And also if using averaging: a LM399 with 7uVpp will spoil a group of references with 3-4uVpp.
Attached noise measurements of LM399#2 (7.2uVpp) and LM399#1 (3.6 uVpp)
y-axis 1 uV/div
x-axis 1 sec/div
And you also want to sort out references with "popcorn" noise
(jumps of several uV or even ppm´s for several seconds or even minutes)
Attached a example of a MAX6350C reference over several minutes with jumps up to 4-5 ppm.
Hi,
Read this and you will see that there is a very low frequency noise component present in even the very best units available:
http://vmetrix.com/ZenerP.pdf
I would not call the seasonal changes as noise.
Since you can correlate it to the environment.
So it is more a drift. (temperature / humidity / pressure dependant).
Noise is something statistical that cannot be predicted.
with best regards
Andreas