IntroductionIn the discussion of
my Metrology Meet 2022 results there were some insightful comments about how some standard resistors might need more than the standard 1s for 10kOhm and 5s for 100 kOhm recommendation that's a common rule-of-thumb for resistance measurements with the HPAK 3458A. To not interfere with the other discussion in that topic I created a separate topic.
This topic has discussed multiple times before, in particular by Dr. Frank's great investigation work:
Precise Offset Compensation Ohm Measurements and Validation of DMMsoptimal configuration of HP 3458A to measure 10kIn short, offset compensation in the 3458A switches on and off the current source during the measurement cycle to measure both the voltage without any current (the offset voltage it's trying to compensate) and the voltage with current applied (the IR drop of the resistor under test). But due to dielectric absorption in the cables and rest of the system, this voltage step might have an effect on the measured voltage even after the default 30 ms delay. This voltage has been noticed as depending on the resistance, with 1 kΩ and below barely needing any delay, 100 kΩ needing the most delay, and above 100 kΩ offset compensation is disabled by the 3458A either way.
The question here was if the SR104 due to its large size and oil tank suffers more strongly from this effect than say a Fluke 5450A or VHP bulk metal foil resistors.
SetupIn these tests I have taken a series of measurements with different delay values, cables and resistors. All measurements were taken by a HP 3458A with ACAL ran at least every 12h, but not while performing a measurement for a specific combination of settings / DUT / cable. All measurements were in 4W Ohms mode, with offset compensation on, 100 NPLC.
I picked my best standard resistors with nominal values of 1 kΩ, 10 kΩ and 100 kΩ:
- HP 11103A 1 kΩ
- ESI SR104 10 kΩ
- Guildline 9330 100 kΩ
- The Fluke 5450A resistance calibrator set to 1 / 10 / 100 kΩ (using the rear binding posts)
I used cables assembled from parts sold by
Forum user ap (Adrian). They are four wire PTFE cables with a woven outer shield terminated in gold-plated copper (low-EMF) binding posts crimped with the proper Knipex crimper and die. I used two cable lengths: 1m (which I used at MM2022) and 2m (which I normally use at home because my 3458A is high above my bench). In all cases the shield was connected to the guard terminal on the 3458A and if available (SR104, 5450A) to the guard terminal on the resistor and the four wires to the force / sense terminals on both sides.
I then had the 3458A take a series of samples starting from DELAY 0 up to the highest value, one measurement per delay value, and then repeat this process at least 10 times. This way I could see the relative change between samples with different delay values and then average the effect over the multiple rounds.
Error bars are standard error of the mean. N >= 9. See for more details the attached CSV file with the results as table. I fit an exponential decay (charging capacitor) model using the Levenberg-Marquardt model weighted by standard error of the mean which I felt made sense and wouldn't overfit. To make sure even the delay=0 point had a finite standard deviation, I used 0.01 (an order of magnitude below the smallest) there.
ResultsHere are the mean relative differences compared to the measurement with DELAY=0 over all rounds I did with the same resistor / value / cable:
It appears that the wire wound resistance standards have much less effect due to dieletric absorption than the 5450A with all its extra wiring and relays. The effect is stronger in the 100 kΩ wire wound Guildline 9330 than in the 10 kΩ SR104, but it's especially strong in the Fluke 5450A when set to 100 kΩ. Based on this 1s for 10 kΩ seems plenty for the SR104, but for the 5450A at least 2s seems necessary, and for 100 kΩ, at least with the 5450A, 5 seconds seems a sensible value.
The results with the 2m cable look quite similar to me. The 5450A-100k measurement has a lower number of rounds compared to most others (only 9), so that explains the larger error bars. The fit is clearly failing for the Fluke 5450A 100 kΩ result.
The 1 kΩ results are on a vertical scale that is a factor 20 magnified, so we're getting down in the noise here.
The results with 2m cable don't look very different from the 1m cable results.
Discussion & conclusionThe fit results are summarized in this table:
| | | V0 | tau (s) | time (s) |
cable | dut_setting | dut | | | |
4w PTFE 1m | 1 kOhm | Fluke 5450A | 0.21 | 5.16e-02 | 0.07 |
| | HP 11103A | -0.13 | 2.14e-01 | 0.20 |
| 10 kOhm | Fluke 5450A | 4.34 | 5.87e-03 | NA |
| | SR104 | 0.50 | 3.82e-01 | 0.88 |
| 100 kOhm | Fluke 5450A | 11.91 | 4.45e-01 | 2.43 |
| | Guildline 9330 | 5.43 | 4.21e-01 | 1.97 |
4w PTFE 2m | 1 kOhm | Fluke 5450A | 0.24 | 5.25e-02 | 0.08 |
| | HP 11103A | -0.12 | 2.95e-01 | 0.27 |
| 10 kOhm | Fluke 5450A | 4.20 | 4.40e-01 | 1.95 |
| | SR104 | 0.77 | 2.37e-01 | 0.65 |
| 100 kOhm | Fluke 5450A | 13.66 | 4.12e-01 | 2.31 |
| | Guildline 9330 | 7.61 | 4.10e-01 | 2.06 |
I calculated time with the condition that the contribution of the dielectric absorption to the uncertainty should be at most 0.05 ppm, which I admit is a somewhat arbitrary number. I used the inverse of the exponential delay function to calculate this time based on the fitting parameters. Obviously the fit for the 5450A 10k 1m measurement did not yield a valid tau. The cable length seems to have a pretty minor effect. The wire wound resistor standards, including the SR104, need less delay than the Fluke 5450A resistance calibrator. I suggest using longer delays when measuring a calibrator like the 5450A than for a bare resistor. I'd summarize the advice as:
- For 1 kΩ and below, use 0.1 s (I don't trust the 11103A results much due to the huge uncertainty, maybe I'll redo those)
- For 10 kΩ with bare resistors, use 1 s
- For 10 kΩ with a calibrator, use 2 s
- For 100 kΩ with bare resistors, use 3 s
- For 100 kΩ with a calibrator, use 5 s (I'm not so convinced by the knee of the fit for the 5450A-100k results)
Note that this is purely based on my cables, resistors, etc, so it may not be applicable to other setups. I'd love to hear feedback on both the data and my analysis of it.
Data as tableMy results are also available as a table as the attachment
measurement-tests-delay-test-results.csv.txt. if you want to look more in depth. I also attached my raw data as a CSV file.