Author Topic: Update of my Keithley 2001  (Read 2342 times)
Offline eurofoxTopic starter
    • Music
« on: June 22, 2018, 01:14:15 pm »
I updated successfully my 7 1/2 digit DMM  to latest firmware and memory 2 extension thanks to information from TiN website.
I calibrate it with 10K, 2VDC and 10VDC, yes it work because with 20K, 2VDC and 20VDC I got a calibration error!
eurofox
Offline TiN
  • Country: ua
    • xDevs.com
Re: Update of my Keithley 2001
« Reply #1 on: June 22, 2018, 01:36:42 pm »
However your DCI/ACV calibration is still likely to be invalid, because those factors require full manufacturing level calibrations with lots more points.
    • Music
« Reply #2 on: June 22, 2018, 02:00:56 pm »
For AC it is not measurable with the source that I have, so it is not relevant and on DCI it is 24µa low on 100ma range.
A full calibration of this instrument cost arms & legs.
eurofox
Offline TiN
  • Country: ua
    • xDevs.com
« Reply #3 on: June 22, 2018, 02:05:03 pm »
Maybe you know someone in your geo who has 5700A/4808 + 2Vrms 1Hz source to do the cal  :)
Online Simon
  • Global Moderator
    • Simon's Electronics
« Reply #4 on: June 22, 2018, 07:18:31 pm »
This entire thread is pointless!