DCC means "Direct Current Comparator", and this is used for current ratios in a resistance bridge, or for current ratios when measuring DC (and up to about 100KHz). See: Danfysik and LEM DCCT's...
I think these results might help someone. Was playing a bit with my LEM Ultrastab IT600-S and calibration equipment while ago.
Also have plan to redo similar test on my home gear, but that will come later on
Brief and non-destructive teardown of these zero-flux current sensors was done year ago and
published here.
Ok, here are the results.
Current source 1: Fluke 5700A
Current source 2: Fluke 5700A + 5205A
Configuration : 7 turns thru LEM
Burden shunt: 2.5 ohm Vishay PG foil 1ppm/C 9W, CSNG
Meter: Keithley 2002 on 200mV range
All gear was freshly calibrated and kept within 2C of cal temp.
0A : 0 mVDC rel
214.1942 uA : 0.00250 mV
2.141942 mA : 0.02505 mA
21.41942 mA : 0.249960 mV
200mA : 2.33431 mV
214.1942mA : 2.49986 mV
1A : 11.6719 mV
2A : 23.3434 mV
2.141942A : 25.00005 mV
High current tests with Fluke 5205A (uncalibrated) + 5700A (calibrated)
1.000000 V(MFC) : 11.64732 mV
2.000000 V(MFC) : 23.31058 mV
2.142857 V(MFC) : 24.98688 mV
With 5205A:
3.000000 V : 34.99193 mV
4.000000 V : 46.66403 mV
5.000000 V : 58.33468 mV
6.000000 V : 70.00847 mV
7.000000 V : 81.68046 mV
8.000000 V : 93.35251 mV
9.000000 V : 105.024558 mV
10.000000 V : 116.69671 mV
11.000000 V : 128.36888 mV
12.000000 V : 140.04104 mV
13.000000 V : 151.71314 mV
14.000000 V : 163.37437 mV
15.000000 V : 175.05746 mV
16.000000 V : 186.72914 mV
16.300000 V : 190.23057 mV
5205A could not go higher, as it's not suitable for large inductive loads and was losing stability after 16A.