If I read your numbers and chart correct (and I have not studied each and every page of the whole thread thourougly, so correct me if I am wrong please), your graph shows a drift of 80ppm downwards within 1khr. This is supposedly after the first upward drift within the first couple hundred hours as shown in figure 4 of the ADR datasheet. So direction seems ok. To do a very rough and ballpark figure sanity check, at an average assumed say 125°C and assuming an Arrhenius Q10 factor of 2 (and there may of course be other drift contributors as well such as the circuitry resistor drifts, die attach...; so this is really a very rough check), and related to the 75°C Tset of the data sheet, that would translate to a drift rate at 75C of about 2,5ppm/khr. Assuming Q10 of 2,5, it would translate to below 1ppm. So the initial drift you have seen so far may actually be correct and not related to "hysteresis" as you state.
The drift direction change is a different story, it will be interesting to see how this develops and what your cycling results are.