Author Topic: Low ohm precision resistor standard and testing  (Read 86243 times)

0 Members and 9 Guests are viewing this topic.

Offline HighVoltageTopic starter

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #100 on: May 23, 2017, 12:40:34 pm »
Thanks gamalot for the clarification, it makes sense now.

Your Hioki low resistance meter looks very good!
They are just sooo expensive.

There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #101 on: May 23, 2017, 06:48:42 pm »
because of the HIOKI RM3545 shown here,  I just updated my old "Resistance Comparison" chart, which I made several years ago, with some of the other gear shown here (and left in the FLUKE 8508A for reference)



Keithley 2460 is not fully shown here, because of the odd accuracy specs <2 Ohm and >20MOhm (which I did not calculate)


edit: found an error in HIOKI RM3545 graph and corrected it
« Last Edit: January 12, 2018, 08:28:29 am by quarks »
 
The following users thanked this post: onemilimeter

Offline HighVoltageTopic starter

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #102 on: May 23, 2017, 07:12:56 pm »
Thanks quarks for this nice chart.
And it becomes very clear now, why the 34470A is not suited for a 1 mOhm measurement as I have tried.

Also a surprise how much better the Fluke 8508A is in comparison to the Keysight 3458A for the full range.

There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline onemilimeter

  • Frequent Contributor
  • **
  • Posts: 314
Re: Low ohm precision resistor standard and testing
« Reply #103 on: May 24, 2017, 05:56:51 am »
And it becomes very clear now, why the 34470A is not suited for a 1 mOhm measurement as I have tried.
Hi... quite sad to read this :(... Is this "the 34470A is not suited for a 1mOhm measurement" mainly because of the low test current 1mA at 100-ohm range? Thanks.
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 963
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #104 on: May 24, 2017, 11:29:07 am »
Today I tried to measure a 1mR Isabellenhütte PBV 0.5% shunt on the 34470A. It showed 0.0010 Ohms.
 

Offline VintageNut

  • Frequent Contributor
  • **
  • Posts: 534
  • Country: 00
Re: Low ohm precision resistor standard and testing
« Reply #105 on: May 24, 2017, 12:30:18 pm »
Today I tried to measure a 1mR Isabellenhütte PBV 0.5% shunt on the 34470A. It showed 0.0010 Ohms.

What is the uncertainty of the measurement according to the datasheet?  You are only using 10 counts of a 10,000 count range.
working instruments :Keithley 260,261,2750,7708, 2000 (calibrated), 2015, 236, 237, 238, 147, 220,  Rigol DG1032  PAR Model 128 Lock-In amplifier, Fluke 332A, Gen Res 4107 KVD, 4107D KVD, Fluke 731B X2 (calibrated), Fluke 5450A (calibrated)
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 963
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #106 on: May 24, 2017, 12:32:48 pm »
Today I tried to measure a 1mR Isabellenhütte PBV 0.5% shunt on the 34470A. It showed 0.0010 Ohms.

What is the uncertainty of the measurement according to the datasheet?  You are only using 10 counts of a 10,000 count range.

I know it is nonesense to measure 1mR directly with the 34470A. I did it only because of the wired measurements HighVoltage took.
 

Offline HighVoltageTopic starter

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #107 on: May 30, 2017, 10:07:37 am »
Today I tried to measure a 1mR Isabellenhütte PBV 0.5% shunt on the 34470A. It showed 0.0010 Ohms.

What is the uncertainty of the measurement according to the datasheet?  You are only using 10 counts of a 10,000 count range.

I know it is nonesense to measure 1mR directly with the 34470A. I did it only because of the wired measurements HighVoltage took.
Thank you for making this measurement on your 34470A
It seems yours is at least better calibrated in this low ohm area than mine and I will have my meter definitely calibrated by Keysight Germany, in the next few weeks.

There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline HighVoltageTopic starter

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #108 on: May 30, 2017, 10:16:07 am »
Does anyone have any experience with the ex ussr milli-ohm standard resistors?

I got my first USSR precision resistor with a 1 Ohm nominal value.
No papers and no info was given by the seller.
So I can only go by what is written on the label

Model P3030 from 1986
It seems the 6 mm studs are pure silver or at least silver plated.

Measurements:
1.000068 Ohm - Keithley 2460 SMU
1.00005 Ohm - Agilent 3458A

I will keep this one hooked up to the Keithley 2460 SMU to take some Data over the next few days.
It will be interesting to see, how stable this one will be in comparison to the Burster resistors.


There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 963
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #109 on: May 30, 2017, 10:36:51 am »
Can you tell me which uncertainty is achievable at 1R with the Keithley SMU?

I think a (or better two) 3458A can achieve 52ppm with an external current source (100mA) within 1 year specifications. Is the SMU better?
 

Offline manganin

  • Regular Contributor
  • *
  • Posts: 196
  • Country: fi
Re: Low ohm precision resistor standard and testing
« Reply #110 on: May 30, 2017, 10:40:55 am »
I got my first USSR precision resistor with a 1 Ohm nominal value. No papers and no info was given by the seller. So I can only go by what is written on the label

Soviet foil technology! Better initial adjustment accurary but less stable than the Krasnodar factory wire wound P321 (which unfortunately also suffers from the Soviet quality: there are very good examples, but you need to buy ten to find one.)

 

Offline HighVoltageTopic starter

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #111 on: May 30, 2017, 12:07:38 pm »
Can you tell me which uncertainty is achievable at 1R with the Keithley SMU?

I think a (or better two) 3458A can achieve 52ppm with an external current source (100mA) within 1 year specifications. Is the SMU better?
I don't think the SMU is better than 52 ppm at 1 Ohm.
If I remember right, the Ohm uncertainty was something like 0.04% + 1 mOhm on the datasheet of the SMU. (I will have to look it up.)

But the 2460 SMU has 1 uOhm resolution in the 2 Ohm range, and the 3458A has 10 uOhm resolution in the 100 Ohm range.

There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline HighVoltageTopic starter

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #112 on: May 30, 2017, 12:11:18 pm »
I got my first USSR precision resistor with a 1 Ohm nominal value. No papers and no info was given by the seller. So I can only go by what is written on the label

Soviet foil technology! Better initial adjustment accurary but less stable than the Krasnodar factory wire wound P321 (which unfortunately also suffers from the Soviet quality: there are very good examples, but you need to buy ten to find one.)

It seems you know these soviet resistors.
Thanks for this explanation.
Is the quality issue found in all ranges of the P321 or more in the lower or upper ranges.
May be I will get a 1 Ohm P321 and compare it to this P3030
There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #113 on: May 30, 2017, 06:08:27 pm »
Can you tell me which uncertainty is achievable at 1R with the Keithley SMU?

I think a (or better two) 3458A can achieve 52ppm with an external current source (100mA) within 1 year specifications. Is the SMU better?

if you look in my comparison chart above, you can see that 2460 accuracy is much worse than 3458
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 963
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #114 on: May 30, 2017, 06:18:35 pm »
Quote
if you look in my comparison chart above, you can see that 2460 accuracy is much worse than 3458

Sorry, but I can't. The specifications for 1R are missing (best volts and amps combination for 1R).

And it seems to me, that you already calculated the best possible resistance uncertainty using volts and amps and not resistance measurement. Is that right? Because, the 3458A has about 10x15ppm+5ppm=155ppm uncertainty at 1R. If one combines 100mA (40ppm) and and 100mV (12ppm) you achieve 52ppm.
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #115 on: May 31, 2017, 12:11:36 pm »
Quote
if you look in my comparison chart above, you can see that 2460 accuracy is much worse than 3458

Sorry, but I can't. The specifications for 1R are missing (best volts and amps combination for 1R).

And it seems to me, that you already calculated the best possible resistance uncertainty using volts and amps and not resistance measurement. Is that right? Because, the 3458A has about 10x15ppm+5ppm=155ppm uncertainty at 1R. If one combines 100mA (40ppm) and and 100mV (12ppm) you achieve 52ppm.

maybe I got your initial question wrong, but for me it is clear that the 3458A is much better than a 2460 for 1 Ohm

in my comparison chart I used resistance 1 year accuracy specs from the datasheets
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 963
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #116 on: May 31, 2017, 12:58:48 pm »
maybe I got your initial question wrong, but for me it is clear that the 3458A is much better than a 2460 for 1 Ohm

It wasn't clear to me, that the Keithley SMU is in no range better than 100ppm. Therefore, 1R is also much worse than the 3458A.
I had a look into the datatsheet now.

in my comparison chart I used resistance 1 year accuracy specs from the datasheets

I read about 50ppm in your chart. The datasheet says 155pmm for 1R and 1year.
 

Offline quarks

  • Frequent Contributor
  • **
  • Posts: 874
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #117 on: May 31, 2017, 02:41:01 pm »
in my comparison chart I used resistance 1 year accuracy specs from the datasheets

I read about 50ppm in your chart. The datasheet says 155pmm for 1R and 1year.

3458A 1 year spec is 15ppm + 5ppm in the 10 Ohm range and for 1 Ohm this makes 65ppm
 

Offline e61_phil

  • Frequent Contributor
  • **
  • Posts: 963
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #118 on: May 31, 2017, 02:48:29 pm »
in my comparison chart I used resistance 1 year accuracy specs from the datasheets

I read about 50ppm in your chart. The datasheet says 155pmm for 1R and 1year.

3458A 1 year spec is 15ppm + 5ppm in the 10 Ohm range and for 1 Ohm this makes 65ppm

Yeyp, sorry. I interchanged ppm of reading and ppm of range.
 

Offline HighVoltageTopic starter

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #119 on: June 01, 2017, 09:56:18 am »
The Russian based Model P3030 1 Ohm resistor was now for almost 48 hours on the Keithley 2460 SMU at 100mA bias current.
And it shows a significant drift of 48 uOhm over 171.000 readings.
But may be that is normal?
And I dont know how much drift is contributed by the SMU.

I have also ordered a Burster 1 Ohm resistor for comparison purpose.



There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 7278
  • Country: hr
Re: Low ohm precision resistor standard and testing
« Reply #120 on: June 02, 2017, 02:58:13 pm »
The Russian based Model P3030 1 Ohm resistor was now for almost 48 hours on the Keithley 2460 SMU at 100mA bias current.
And it shows a significant drift of 48 uOhm over 171.000 readings.
But may be that is normal?
And I dont know how much drift is contributed by the SMU.

I have also ordered a Burster 1 Ohm resistor for comparison purpose.

Well,
it looks like you have P-P 48 ppm drift+noise on 1 Ohm..
Noise is 9.686 ppm RMS, roughly it looks like about 20 ppm P-P.. So median drifted about 20 ppm.
Keithley 2460 has Extended specs +-40 ppm/°C of reading.... (0,1% /°C of +-0,04%)... on 20 Ohm/100mA range..
Keithley alone could have drifted that much with change in temp of less than 0.5°C.

But that noise looks high.. 100mA range on 2460 is specified at 100nA RMS noise..
So contribution to noise from SMU should be in 100nV range RMS on 1Ohm resistor.. You seem to have 100x more noise..

Maybe cabling, external disturbances?
Did I calculate wrong?

Regards,

Sinisa
 

Offline HighVoltageTopic starter

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #121 on: June 03, 2017, 01:28:37 pm »
Maybe cabling, external disturbances?
Regards,

Sinisa

I also suspected the cable and made a new one.

Well, I also got a new resistor to play with.
This is a Russian P321 1 Ohm in good condition, was suppose to be "new".
But it came without any paperwork.

I made a cable, Silver plated copper with crimped lugs (Low EMF) and PTFE insulation.
The shield is grounded to the 2460 SMU

The results are much better already, with close to 100.000 readings.
Pk/Pk is only 26 uOhm.


There are 3 kinds of people in this world, those who can count and those who can not.
 
The following users thanked this post: 2N3055

Offline HighVoltageTopic starter

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #122 on: July 05, 2017, 02:16:02 pm »
Finally I got a Burster 1240-1, a 1 Ohm calibration resistor, "Made in Germany"
Condition: looks like new, came in original box.
Thank you eBay !

This one was made on 12.05.2000
Has a listed tolerance of 0.025% and a stated Imax of 1.5A
I have tested this resistor on my Keithley 2460 with 100 mA for a few hours.

Here are the results:
- Measured average value: 1.000 081 Ohm
- Pk to Pk over 6841 readings: 19.6 uOhm
I think I can be very happy with this resistor and first results.


There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline alm

  • Super Contributor
  • ***
  • Posts: 2903
  • Country: 00
Re: Low ohm precision resistor standard and testing
« Reply #123 on: July 05, 2017, 02:35:20 pm »
Noise for the Burster looks very similar to the P321. Still, the 3 uVrms noise seems high compared to the SMU specifications, suggesting that improvements are still possible with better shielding.

I think it is hard to make any statements about short-term drift since it appears to be buried in the noise of the setup. Averaging might help, or just measure over larger intervals.

Offline HighVoltageTopic starter

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Low ohm precision resistor standard and testing
« Reply #124 on: July 06, 2017, 10:00:37 am »
I think it is hard to make any statements about short-term drift since it appears to be buried in the noise of the setup. Averaging might help, or just measure over larger intervals.

OK, I turned the "Filter" ON and had it run over night.
Now we can see a drift of 34 uOhm
The temperature did fall about 1.5 °C in the same time.

May be it is time for me to get a temperature chamber.
There are 3 kinds of people in this world, those who can count and those who can not.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf