IntroductionWhen putting Zener diodes in circuit, individually determining the "right" current to match the zero-TC with the intended "temperature of use" is of great value. With a fixture holding only the Zener diode, the thermal coupling is a lot tighter and the sweep and equilibrium times can be a lot faster, compared to a whole reference pcb in an air chamber. Even if the final current needs to be tweaked, to correct for the tempco of a whole assembly, the information collected can also be of use for that.
Part 1 - HardwareI started by making a construction in Freecad to get an idea of the dimensions. The basic idea was an aluminum clamp for the TO3 case, coupled to a Peltier module, which is in turn coupled to a heatsink. The construction was of great use later, while fabricating the parts. All machining could be done with a drill press, table saw and hand tools - no CNC necessary.
The clamp was made from 8mm aluminum plate. The PT100 for the temperature feedback to the TCB-NE Peltier controller is a Heraeus M310B. It is thermally coupled to the aluminum block with boron nitride thermal paste - not needed for the temperature range, but since it dries to a "crumply plaster" like consistency, removing the sensor is a lot less messy than with regular thermal compound.
The lid to compress the stack onto the heatsink was made of FR2 (Pertinax) and had strips glued onto the bottom to center the clamp on the Peltier. It also acts as a spring to allow for thermal expansion. The FR2 was deemed not thermally conductive enough to be a problem.
The thermal interfaces were coated in normal white thermal grease and the stack assembled. The insulation is made out of layers of hot wire cut Styrofoam.
The lid plate is pressed onto the heatsink with metal screws. The small PCB on the fan is an Vbe multiplier, to adjust the fan speed by series voltage drop. This allows it to be run from the same power supply as the Peltier module controller. The heatsink doesn't get hot enough to warrant a thermally controlled fan, slight air movement is enough.
Soon after completion, the realization came that on a dual channel SMU, a dual fixture would obviously make sense. The dual clamp is a mirrored version of the original, just the temperature sensor was moved to the center and because of that, the cable exits in a less convenient position. The clamp is one piece and as wide as the Peltier module. On the new lid, the glued centering strips were omitted, since it was found that friction was enough to keep the clamp centered.
This is how it looks in use, with kelvin clips to get a reading right at the component leads. With the 12V / 4A Peltier module the temperature can be changed from 55°C to 15°C (131°F to 59°F) in slightly above 1 minute.
I didn't attach the Freecad files, since you would need to find the same random heatsink and fan, which seems pretty unlikely. The detailed dimensions of the clamp are not critical aside from the center bore - I found TO3 diameter variation (between manufacturers) to be bigger than the clamping adjustment range and had to put in a strip of copper foil to fill the gap.