Hallo, I´m asking for advice, if there is room for Improvement.
The voltage source is made to deliver input signal (SMA-Connector, terminated with 50Ohm, input voltage 100mV..5V) to a laser designed for microscopy.
To control laser during temperatur chamber test which measure stability of optical power.
I choose the components to later build a version with 100mV / 5V / adjustable 0-5V that can deliver 100mA at 5V into the 50Ohm
Driver: LT3040 low noise, high PSRR precision buffer, eval board DC2783A used as base, operated at +12V
Reference: MAX6350 (I used this for 5V, reasonable low drift and low noise - in combination with the Trim option to compensate for the cable loss caused by the 100mA at 5V) with 3 Capatitors (In, out, noise reject)
I did glue/solder the Ref to the Evalboard, Output fed to a divider 4,99k 0.01% 5ppm, 95R3 0.1% 25ppm and a 10R Trimmer.
Adjusting it using T-piece at laser input to DMM Keysight 34461A, i ca adjust to +/-5µV to 100mV display, so the accuracy is mainly determined by DMM.
Operating it with a channel of a HMC8043 Powersupply i got noticable ripple, so I added a 20mH common mode choke.
Noise now measured with scope R&S RTA4004 (500µV/Div, 10ms/Div, 20MHz) to 30µV, (25µ measured with 50Ohm terminator connected), so corrected sqrt(30^2-25^2)= 16,6µV noise.
Plan is now to mount into metal box, with isolated BNC Socket, box connected to board ground at point.
Any suggestions to the internal / external wiring, is very susceptible, probably triboelectricity, microphonics of capacitors is so far minor problem.