Author Topic: Keithley 2612 teardown and repair  (Read 15881 times)
Offline dzseki
  • Country: hu
Re: Keithley 2612 teardown and repair
« Reply #25 on: September 11, 2020, 11:21:33 am »
Anyone here on eevblog that has a 2612 or similar?
At work we have a few depending on their age we have 2612 no suffix, A and B, we even have 2635B and 2536B.
Ocassionaly they fail a channel...
HP 1720A scope with HP 1120A probe, EMG 12563 pulse generator, EMG 1257 function generator, EMG 1172B signal generator, MEV TR-1660C bench multimeter
Offline MegaVoltTopic starter
  • Country: by
« Reply #26 on: September 14, 2020, 08:31:58 am »
Let's continue the repair :)
I could not save the PCB :( Unfortunately, there was a short circuit inside and all the burnt part had to be removed :(
Here are photos of the result. And the circuit of the primary winding itself.
I have to restore it.
  • Country: by
« Reply #27 on: September 17, 2020, 08:27:02 am »
I restored the primary winding of the transformer and the device started working !!!! It turns on without a single error. It was already late and I did not test it further. Plus you need to disable interlock.
But I found that two LM2940C LDO stabilizers are very hot. (you can't hold your finger.). And I don't understand yet, is it another kind of error on the board, or it was so laid down during development. It feels like they have 24V at the input :)) And at the output 5V. And 50mA current causes 1W loss .... Maybe someone knows more?
Offline razvan784
« Reply #28 on: September 17, 2020, 09:19:00 am »
They may rely on airflow from the fan for additional cooling. With the case open I imagine the airflow is significantly lower.
  • Country: by
« Reply #29 on: September 17, 2020, 09:23:31 am »
The fan has been turned off. Perhaps with the lid closed, the temperature will rise and it will work ... But it seems to me that the fan monitors the temperature of the radiators of powerful transistors.
  • Country: by
« Reply #30 on: September 18, 2020, 08:11:24 am »
I checked the input voltage. At the input 9V. The output is 5 volts. The voltage is in the isolated area to power one SMU channel.
It looks like it was intended. And I'm really thinking about radiators for these LDOs.
Offline Kleinstein
« Reply #31 on: September 18, 2020, 11:00:00 am »
Some 4 V drop at 50 mA is not that bad, to really need a radiator. The regulator closer to the edge of the board has an extra isolating cover - so the radiator might need isolation too.
The special very loose wound isolation transformer looks funny. It it would be about low capacitance they should have added another spacer on top of the core.
  • Country: by
« Reply #32 on: September 18, 2020, 11:08:10 am »
50 mA was my guess based on the estimated power. In reality, the voltage is less. The current means more. I am assuming 250 mA. But I did not measure this figure. And I don't know what consumption should be.
To be honest, I do not understand why there is additional insulation that worsens convection. The body of the device is far enough away.
Maybe there is a picture of how it should be?
The following users thanked this post: Nebulex
  • Country: by
« Reply #33 on: October 05, 2020, 03:16:31 pm »
Cables have come. Here's what they are inside:
Offline Nebulex
« Reply #34 on: November 24, 2020, 07:27:01 am »
Hello I think I have the same device with a different error. If you like I could take some pictures of the part you want.
The following users thanked this post: MegaVolt
  • Country: by
« Reply #35 on: January 11, 2021, 02:02:39 pm »
I am finally back and can continue :)
Add radiators for very hot stabilizers:
Offline Davidwasalreadytaken
« Reply #36 on: August 17, 2023, 04:37:39 pm »
I stumbled over this thread and I wonder if you ever got this device working again
  • Country: by
« Reply #37 on: August 17, 2023, 06:47:59 pm »
Yes, sure. The device is working. Here I wrote about it.
https://www.eevblog.com/forum/metrology/keythley-2612-teardown-and-repair/msg3236456/#msg3236456
Once I used it all day as a bipolar power source :))))) and by the end of the day there was a bad smell :( You need to disassemble and carefully study where it still heats up more than necessary.
Online ch_scr
« Reply #38 on: August 17, 2023, 07:02:19 pm »
One has to add, even fully working, these devices (K2600 series SMU) don't get warm, they get hot. Even doing nothing, just sitting turned on. If you have one, observe the (quite wild IMHO) warmup drift (i.e. measuring an external stable voltage). I got the impression, they use the idle power dissipation as a heater with the whole device as an oven so to speak, and the fan to regulate it, to stay at an (constant) elevated temperature.