Author Topic: Keysight Staandards Lab Fluke 732B calibration  (Read 7070 times)

0 Members and 2 Guests are viewing this topic.

Offline alligatorbluesTopic starter

  • Regular Contributor
  • *
  • Posts: 155
  • Country: us
Keysight Staandards Lab Fluke 732B calibration
« on: June 16, 2023, 02:30:16 am »
Finally scanned pdf's and attached,

I sent in my master 732B to Keysight for a direct comparison to its JJ array. It's a fantastic deal, $1,000.00 cheaper than NIST. The output value was based on 217 readings with a total uncertainty of 20.5 nV. And for the $1,600.00 it charges, you also get the 1.018 output checked. Mine had an uncertainty of 15.2 nV.

The output value is given to 10 digits. And the monitoring took 2 weeks. I have some previous cal points, so now I can estimate the output on a daily basis. After 3 years, my trending figure was within 0.3 ppm. So, I should be able to do better now.
« Last Edit: December 13, 2023, 05:12:39 am by alligatorblues »
 
The following users thanked this post: Rax, mendip_discovery, Okertime

Offline mendip_discovery

  • Frequent Contributor
  • **
  • Posts: 998
  • Country: gb
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #1 on: June 16, 2023, 12:02:54 pm »
Still makes me wince at the cost of it.
Motorcyclist, Nerd, and I work in a Calibration Lab :-)
--
So everyone is clear, Calibration = Taking Measurement against a known source, Verification = Checking Calibration against Specification, Adjustment = Adjusting the unit to be within specifications.
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #2 on: June 16, 2023, 03:36:22 pm »
I sent in my master 732B to Keysight for a direct comparison to its JJ array. It's a fantastic deal, $1,000.00 cheaper than NIST. The output value was based on 217 readings with a total uncertainty of 20.5 nV. And for the $1,600.00 it charges, you also get the 1.018 output checked. Mine had an uncertainty of 15.2 nV.

The output value is given to 10 digits. And the monitoring took 2 weeks. I have some previous cal points, so now I can estimate the output on a daily basis. After 3 years, my trending figure was within 0.3 ppm. So, I should be able to do better now.


Did you send the 732B from Italy to Keysight Colorado / USA?
Why did you not use the FLUKE service in Europe?
Can yo share a scan of the calibration certificate? (Remove your name ...)
There are 3 kinds of people in this world, those who can count and those who can not.
 
The following users thanked this post: TiN

Offline alligatorbluesTopic starter

  • Regular Contributor
  • *
  • Posts: 155
  • Country: us
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #3 on: September 14, 2023, 05:17:19 pm »
My flag is wrong. I'M in the USA. Sure, I'll post the cert.
 

Offline alligatorbluesTopic starter

  • Regular Contributor
  • *
  • Posts: 155
  • Country: us
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #4 on: November 10, 2023, 04:44:12 pm »
The cal docs are still sitting in the scanner ADF. It won't work with linux. It's an Ethernet connected Xerox B215 MFP. When I figure it out,, I'll post the scans.
 
The following users thanked this post: splin, Okertime

Offline donlisms

  • Frequent Contributor
  • **
  • Posts: 302
  • Country: us
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #5 on: November 10, 2023, 05:27:51 pm »
Computing defies calibration.  :)

I'm looking forward to whatever you can share about this interesting experience. Thanks for this so far!
 

Offline alligatorbluesTopic starter

  • Regular Contributor
  • *
  • Posts: 155
  • Country: us
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #6 on: November 11, 2023, 06:04:26 pm »
Working with Keysigut on purchasing laborarory services from it, proved d9fficult at first. Keysight has detailed and inflexible methods to serve its customers. It does not sell certain things to the general public, and the stqndards lab cal is one.  I don't have a company website, domain, email qddress, Dunn and Bradstreet number and rating, company checking account, street address in a commercially zoned area, company phone #, or fax number; so Keysight wanted proof that my company name in my profile in my website account, was actually located at the corresponding address. .

I sent as proof of company address, a photo of an envelope with the company name and address, which is my home address. It didn't fly. So Keysight made my company name my actual name, so I could order as a business according to Keysight's policies. Keysight is such a large company, with 0ver 100 locations, and thousands of products available worldwide, along with facilities to do service and repair, that unless you have experience  dealing with it, it can take a long time..

For the standards lab calibration aqainst  the JJ array, there are at least 10 variations depending on the neeeds of the customer. i have traceability, anab logo, 17025 and Z540.
 

Offline RandallMcRee

  • Frequent Contributor
  • **
  • Posts: 542
  • Country: us
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #7 on: November 11, 2023, 10:10:45 pm »

Hmmm. Just take a picture of the page with your phone. No one needs a real scan! (Just like in the 1980s cover the top secrety stuff before the photo is taken).

Thanks,
Randall
 
The following users thanked this post: aronake

Offline Laurent G

  • Newbie
  • Posts: 3
  • Country: fr
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #8 on: November 30, 2023, 08:27:09 pm »
I'm interresting too for watching the certificate, please
 

Offline alligatorbluesTopic starter

  • Regular Contributor
  • *
  • Posts: 155
  • Country: us
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #9 on: December 13, 2023, 05:16:46 am »
See OP for pdf's of cert.
 

Offline MegaVolt

  • Frequent Contributor
  • **
  • Posts: 930
  • Country: by
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #10 on: December 13, 2023, 11:15:31 am »
See OP for pdf's of cert.
Can I have a copy too?
 

Offline guenthert

  • Frequent Contributor
  • **
  • Posts: 761
  • Country: de
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #11 on: December 15, 2023, 01:35:14 pm »
"This Zenner voltage standard"?  I don't want to be that guy, but for the fee they charge, I'd think they could have a 2nd person have a glance at the template, if not the actual certificate.  I mean, that's the kind of error, the customer can identify ...

I wonder, whether the certificate comes with additional information on how the uncertainties are calculated.
« Last Edit: December 15, 2023, 01:57:49 pm by guenthert »
 

Offline alligatorbluesTopic starter

  • Regular Contributor
  • *
  • Posts: 155
  • Country: us
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #12 on: December 22, 2023, 03:45:39 am »
I wonder, whether the certificate comes with additional information on how the uncertainties are calculated.
It says uncertainties are according ISO GUM manual.
 

Offline alligatorbluesTopic starter

  • Regular Contributor
  • *
  • Posts: 155
  • Country: us
Re: Keysight Staandards Lab Fluke 732B calibration
« Reply #13 on: December 22, 2023, 03:47:46 am »
See OP for pdf's of cert.
Can I have a copy too?
I can't stop you. But if it's all the same, I'd prefer the documents be viewed on the forum.
 
The following users thanked this post: MegaVolt