Have you compared specs of 34410 against 2450? My guess is that voltage is not a big deal but current will be a big deal.
The 2450 is optimized for quiet low current source and measure.
I would be surprised if you can calibrate the 2450 low current range with only a 34410.
I think you need a very good voltage source and very good high resistance standards and possibly a picoammeter or electrometer.
working instruments :Keithley 260,261,2750,7708, 2000 (calibrated), 2015, 236, 237, 238, 147, 220,  Rigol DG1032  PAR Model 128 Lock-In amplifier, Fluke 332A, Gen Res 4107 KVD, 4107D KVD, Fluke 731B X2 (calibrated), Fluke 5450A (calibrated)
Offline VintageNut
« Reply #3 on: June 24, 2020, 11:38:41 pm »
From my experience, the lowest current source and measure ranges drift the most and before the higher current ranges.