One more thing to do before alignments. You need to remove the optical unit from the chassis, then remove the nut and washer on the stud on the lamp end, remove the 1st shield.
Remove the internal nut on the stud and remove the 2nd shield.
Remove the 3rd shield. Remove the 3 tiny phillips around the lamp assy. circumference and pull the lamp assy out.
Locate and inspect the 1.33K resistor. If it is original it will be a film resistor and will need to be replaced. (even if it looks good replace it!) They degrade and crack over time! I replace them with a 1.3K 2% MOX resistor. See PIX.
![Inspecting, aligning, and testing an HP 5065A Rubidium](https://www.eevblog.com/forum/index.php?action=dlattach;topic=248240.0;attach=1033964;image)
If careful you can do it without further disassembly but you can remove the 3 big phillips on the same side as the 3 tiny ones and it will come apart. Be careful not to bump the lamp or lose the 3 spacers that fall out.
This unit also had a cracked 12 Ohm that I replaced.
Reassemble everything making sure that you space the inside nut so that the outer shield will be snugly clamped between it and the outer washer and nut!
Reinstall the optical unit into the chassis.
Now for some alignments! Warmed up and in open loop for now.
First a match of the 137HZ filter frequency of the A7 module is made.
I use a short wide bladed tuning tool cut down to 5/8" with heat shrink on it to give better purchase. This allows tuning the card in place without an extender.
Clip an oscilloscope to Pin 13 on the A8 edge connector. You will see a Yellow wire connected to this pin.
Adjust the Quartz fine frequency to give you an observable signal on the scope.
Now adjust the R8 pot on the A8 module for a peak amplitude.
You have just adjusted the A8 output frequency to match the "as built" center frequency of the A7 fundamental filter. This will make the phasing adjustment
we will make later much easier!
Move the scope to A7TP2 and adjust the fine Quartz pot for zero on the error
meter.
On A3 Set the core of L22 or L25 to a little over 1/8" from the top of the form and then leave it
there.
Observing the TP2 signal adjust for a peak on A3 the matching network C61, C55 or C59 and bias pot R40 or R53. Depending on which A3 version you have.
Check TP2 on A3 to make sure it reads >4.0VDC. If not try a different position of the R40 or R53 pot.
Installment 4 to follow!
Cheers,
Corby