Author Topic: Influence of resistors in LM399 reference circuit?  (Read 16317 times)
Offline carl_labTopic starter
« on: May 19, 2017, 10:44:37 am »
I've separated LM399 topic from thread "Resistor Set for LTZ1000 positive standard 7V circuit"
because it was off topic:
#14
...R1 is not critical[/b], also not its absolute value.
Yes, it is not as critical, as given in the LTZ1000 datasheet.
This has been proven by several volt-nuts consistently.
The real, measured attenuation factors for drifts are about +/-90 for R4/R5, -600..-800 for R1, -250 for R2, and about -1000 for R3.
Thanks for this valuable information.
Is there comparable info for LM399 application?
The LM399 is not as adjustable, there is active circuitry limiting the current through the zener, and the on-chip temperature is not adjustable either, so any more than about 1mA through the zener is wasted and it always runs at about 90C.
Thanks, I did not mean temperature setting resistor, but influence of the other resistors to standard self biasig LM399 application.
Hi Carl,
If you arrange it so that you use a JFET at the output of the op-amp, and set the OP-amp with resistor scaling so that it outputs 10V then you only have 3 WW resistors in the whole reference. But those all need to be quite stable. You would have about 3K as a current setting resistor, the zener has a dynamic impedance of about 1 ohm, so 100ppm shift should change the output by about 0.3ppm, and for the resistance scaling that could be calculated 3k/7k for instance as scaling resistors that is about 1:1 change in value to change in output so they need to be stable.
Thanks, MK.
@Phil
I should have been more precise, I meant the resistors in a self biasing circuit simillar to this one (without variable trimming resistor):
(sorry for boarding this "LTZ1000 7V circuit" thread)
If you put the ouput of the op-amp to the gate of a FET then you can do away with the 200k startup resistor.
I don't see the difference at startup with or without using a buffering FET at OP output.
Without startup resistor startup voltage at OP's pin 3 is 0V -> output voltage of OP is 0V -> no chance to power up reference only by feedback (5k) resistor.
Am I wrong?
In 34401A DMM Agilent uses a diode in series to startup pull-up resistor to 5V, so, when reference voltage rises to 7V (by feedback resistor), the diode is reverse biased and startup current will decrease to very low leakage diode current, so startup resistor and its stability is irrelevant... (I think...)
« Last Edit: May 19, 2017, 10:55:53 am by carl_lab »
Offline Gyro
Re: Influence of resistors in LM399 reference circuit?
« Reply #1 on: May 19, 2017, 11:20:59 am »
You could use a resistive divider from the supply and a diode rather than a single resistor. That would avoid needing a 5V supply and still take the startup out of circuit for normal operation.
EDIT: If you set the divider to output just a little below the zener voltage then you should get very low leakage through the diode.
« Last Edit: May 19, 2017, 11:26:35 am by Gyro »
Best Regards, Chris
« Reply #2 on: May 19, 2017, 11:25:07 am »
That's a good idea, thank you!
BTW, that's the 34401A reference circuit:
Why did Agilent set Iz = 2mA?
« Last Edit: May 19, 2017, 12:42:38 pm by carl_lab »
The following users thanked this post: cellularmitosis, Rafael
Offline MK
« Reply #3 on: May 19, 2017, 11:30:29 am »
Hi Carl,
The reason that a JFET gives you a self starting circuit is to look at the VGS(off) of the BF245 for example:
V DS  = 15V, I D = 200µA
BF245A         -0.4      -2.2
BF245B         -1.6      -3.8
BF245C         -3.2      -7.5      IDSS= 12-25mA
that means that at startup when the output of the opamp is at zero then the bottom of the BF245C fet channel is between +3.2 and +7.5 Volts, more than enough to force the inputs into the common mode range, so it will always self start, and the output is mostly protected against shorts ( the FET may overheat). It also keeps the opamp from drifting due to self heating.
« Reply #4 on: May 19, 2017, 12:26:48 pm »
Thanks for explanation.
So this is what you meant?
What about additional noise of the FET?
« Last Edit: May 19, 2017, 12:42:23 pm by carl_lab »
Offline Kleinstein
« Reply #5 on: May 19, 2017, 12:42:50 pm »
The JFET is inside the feedback loop of the OP. So the noise of the JFET is compensated by the OP and thus not relevant. Anyway the LM399 is way more noisy. So one might consider noise filtering (e.g. 5 K and 2 µF) between the reference and the non inverting input of the OP. This could at least attenuate some of the higher frequency noise.
A 2 mA reference current is quite high, and does not really help. However the LM399 runs hot anyway, so some additional heating from the reference current does not really matter.
The following users thanked this post: carl_lab
Offline MK
« Reply #6 on: May 19, 2017, 01:54:54 pm »
As Kleinstein says, the R409 could be 2K5, 2K7, or 3K as any extra current is wasted in the shunt transistor internal to the "Zener".
Also thinking about fault conditions you would want a 10K or so between the Op-amp and the Fet to reduce the current through the gate during a sustained short on the output, it could even be a carbon composition resistor.
The noise is less than the Zener so can be ignored, if it was a 2DW234 then one might want a quieter FET perhaps.
Offline Kleinstein
« Reply #7 on: May 19, 2017, 03:16:11 pm »
Even with a noiseless reference, the noise of the JFET does not matter. It is inside the OPs loops, so it would be attenuated by the OPs open loop gain. So one would only notice noise from the FET at high frequencies like 1 MHz and up. There is more noise from the 12.85 K  resistor. In the lower frequency range the noise is mainly from the LM399.
A series resistor is a good idea an it would not hurt, if not too large.
Keep in mind that the shown circuit can start to oscillate with capacitive load. If driving a capacitive load C441 should be removed and a capacitor from the OPs output to inverting input.
Offline tszaboo
« Reply #8 on: May 19, 2017, 03:42:14 pm »
Volt nutting aside: Can someone explain what is the point of the LM399 in the 21 century? You just replace the problem "I need a stable 10V reference" with " I need a precision thermally coupled matched pair of resistors". That is always the big issue for me. I mean, I can just buy a 10V reference, with guaranteed 2ppm/K tempco. Is it easier to buy 1.5 ppm tracking tempco resistors? Unless I'm mistaking, you either need to make your own network from something like a Vishay foil resistors, and then hope for the best, or order custom networks. Both of which are going to cost a lot.
And I'm not seeing any projects, where they would use the magic Datron 4910 circuit where you actually dont need high precision dividers.
Am I missing something here?
Offline Kleinstein
« Reply #9 on: May 19, 2017, 05:03:25 pm »
For getting a stable 10 V, there are better/competing ways than the LM399. Even if modern references specify a low TC, there might be board stress, humidity and hysteresis issues that add, if the reference is not in an hermetic metal case with long legs. Not many modern refs in the old style metal case.
However the LM399 us still good if you need a stable 7,xx V reference. There are also alternative ways (e.g. charge pump)  to go from 7 V to 10 V or today more likely to go from 7 V to about 5 V, or maybe 3.5 V.
« Reply #10 on: May 19, 2017, 06:13:28 pm »
Maybe I can get a few dumped K2000 and/or 34401A - probably not repairable.
I think about cannibalizing them (LM399 and precision resistors/networks) to make a 10V/7V reference, for calibration of max. 5,5 digits DMMs or for some experimenting. It does not need to have LTZ1000 stability.
« Last Edit: May 19, 2017, 06:18:49 pm by carl_lab »
Offline Mickle T.
  • Country: ru
« Reply #11 on: May 19, 2017, 07:10:28 pm »
Over the past few years I've seen a tons of volt-nuts projects with a PWM and charge-pump conveters for the LTZ (but mostly - Chinese projects).
Offline David Hess
  • DavidH
« Reply #12 on: May 20, 2017, 07:11:51 pm »
Zener references are lower noise than bandgap references.
Offline Echo88
« Reply #13 on: May 20, 2017, 08:11:22 pm »
@ Mickle-T: Can you point us to those chinese LTZ-charge-pump-circuits?
Offline cellularmitosis
« Reply #14 on: May 21, 2017, 08:05:06 am »
I was interested in the question of resistor influence on the LM399 circuit presented in the LM399 thread.
I wrote a little python script to figure this out:  https://github.com/pepaslabs/small-lm399/blob/master/scripts/lm399.py
The script's "solver" is based on the idea that Vz changes by 1uV for every deviation of 1uA of zener current from the nominal 1mA.
(Caveat: I have no idea what I'm doing and have not verified any of this in spice!)
Here's an example of running the script.  Let's use Rf=36k, Rg=82k, Rz=3k, and assume an LM399 with Vz=6.95V@1mA.
$ ./lm399.py rf=36000 rg=82000 rz=3000 vz=6.95 vos=0
using rg: 82000.0
using rf: 36000.0
using rz: 3000.0
using vz_nom: 6.95
using vos: 0.0
solving...
delta_vz: 0.0000170731707314
delta_vz: 0.0000000024985134
delta_vz: 0.0000000000003650
delta_vz: 0.0000000000000000
irf:            0.084,756 mA
vopout:         10.001,244,085 V
iz:             1.017,076 mA
vz:             6.950,017,076 V
Here's what I was able to surmise.  (I'm relatively new to ppm calculations, please check my math!)
  • Changing Rf from 36k to 36.036k changes Vz from 6.950,017,076V to 6.950,018,093V.
  • I.e. increasing Rf by 0.1% (1,000ppm) increases Vz by 1.017uV (0.146ppm)
  • Changing Rg from 82k to 82.082k changes Vz from 6.950,017,076V to 6.950,016,059V.
  • I.e. increasing Rg by 0.1% (1,000ppm) decreases Vz by 1.017uV (0.146ppm)
  • Changing Rz from 3k to 3.003k changes Vz from 6.950,017,076V to 6.950,016,059V.
  • I.e. increasing Rz by 0.1% (1,000ppm) decreases Vz by 1.017uV (0.146ppm)
So, all three resistors have the same influence on Vz.
If you were using 1% metal film resistors with 100ppm/C, as long as your lab stayed within +/- 5C they'd only cause 1uV of Vz drift.
(Of course, if Rf and Rg drift in the same direction, there is no change to Vz!)
  • Changing Vos from 0 to 1mV changes Vz from 6.950,017,076V to 6.950,016,929V.
  • I.e. increasing Vos by 1mV decreases Vz by 147nV (0.0212ppm)
For reference, an LM358 sees a Vos drift of 7uV per C, so 1mV of Vos drift would require a 143C swing.  To get this drift to show up on a 6.5-digit meter (10uV resolution), you'd need to swing 9,724C :)
Now, the above is just the impact on Vz.  "Vopout" (the 10V output of the opamp) is considerably more sensitive.
  • Changing Rf from 36k to 36.036k changes Vopout from 10.001,244,085V to 10.004,296,776V.
  • I.e. increasing Rf by 0.1% (1,000ppm) increases Vopout by 3.053mV (305.2ppm)
  • Changing Rg from 82k to 82.082k changes Vopout from 10.001,244,085V to 9.998,194,444 V.
  • I.e. increasing Rg by 0.1% (1,000ppm) decreases Vopout by 3.0496mV (304.9ppm)
  • Changing Rz from 3k to 3.003k changes Vopout from 10.001,244,085V to 10.001,242,622V.
  • I.e. increasing Rz by 0.1% (1,000ppm) decreases Vopout by 1.463uV (0.146ppm)
So Rf and Rg are very influential on Vopout, but Rz is not (thus NANDblog's comment about needing a pair of unobtainium resistors?).
If you were using Vishay metal foil resistors, the same +/-5C swing would be a 20ppm swing, which means Vopout would swing by ~6ppm (60uV, or 6 counts on a 6.5-digit meter).
  • Changing Vos from 0 to 1mV changes Vopout from 10.001,244,085V to 9.999,804,849V.
  • I.e. increasing Vos by 1mV decreases Vopout by 1.439mV (143.8ppm)
Let's examine a more realistic Vos drift of 10uV:
  • Changing Vos from 0 to 10uV changes Vopout from 10.001,244,085V to 10.001,229,692V.
  • I.e. increasing Vos by 10uV decreases Vopout by 14.393uV (1.438ppm)
So, creating a 10V reference is much harder than creating a 7V reference!
« Last Edit: May 21, 2017, 08:09:15 am by cellularmitosis »
LTZs: KX FX MX CX PX Frank A9 QX
The following users thanked this post: Rafael
« Reply #15 on: May 21, 2017, 08:23:06 am »
(for reference, here's the "portable calibrator" circuit rearranged in the same style.  The only difference is the addition of a 200k resistor, which I assume ensures startup).
« Reply #16 on: May 21, 2017, 08:56:06 am »
Here's all of that summarized in a table, with very heavily rounded figures for ease of digestion.
Basically, the 10V output is 2,000x driftier than the 7V output.
« Last Edit: May 21, 2017, 09:01:48 am by cellularmitosis »
Offline RandallMcRee
« Reply #17 on: May 21, 2017, 04:46:54 pm »
Hey cellular--
Have you seen this analysis?
http://satcom.tonnarelli.com/files/LM299/Walter_G_Jung.pdf
According to Jung the gain resistors only affect the scaled output, so of course resistor drift there does affect the scaled output, only. It would be wrong, of course, to conclude that the 7V output is less drifty because you have not taken into account the zener itself. Also that url is just a snippet of a chapter which I do not have access to but I believe Jung is then going to shoot holes in your theory that opamp Vos is negligible!  8) There *is* most likely a reason that a precision opamp is recommended there.
Thanks,
Randy
« Reply #18 on: May 21, 2017, 05:59:13 pm »
Oh, thanks I'll definitely take a look!
You're right though -- I wasn't including the drift of the LM399 -- I should have said "the resistor influence is 2000x driftier on the 10V output"
Offline Andreas
« Reply #19 on: May 21, 2017, 06:41:50 pm »
Hello,
I would not worry about the T.C. of the LM399 near room temperature.
On my references it is usually well below 20uV change over a 30 deg C range.
https://www.eevblog.com/forum/metrology/lm399-based-10-v-reference/msg1119139/#msg1119139
with best regards
Andreas
« Reply #20 on: May 22, 2017, 02:32:19 am »
Andreas, thanks for the data, I always look forward to seeing your plots!  Your thorough empirical approach is very inspiring to budding volt nuts :)
« Reply #21 on: May 22, 2017, 06:20:20 am »
Thanks, CM for elaboration.
That's about the same, I roughly calculated.
BTW: Differential resistance of LM399 zener is about 0.7mV/mA (=Ohm):
« Last Edit: May 22, 2017, 12:45:00 pm by carl_lab »
« Reply #22 on: May 23, 2017, 04:39:48 am »
Thanks for posting your calculation results!
Offline tszaboo
« Reply #23 on: May 24, 2017, 08:51:02 am »
Hey cellular--
Thanks,
Randy
Of course it is negligible. Even an oldstyle opamp, like an OP07 has max 1.3 uV/K offset drift, which is amplified, A<1.5, so lets calculate with 3uV/K. That is 0.2ppm worst case. There are much better opamps, and the typical characteristics will be better.
All these calculations of ppm change/ppm output hurts my head. You are describing dimensionless parameters. The correct way of saying it, is simple. You are calculating the sensitivity of the circuit for several parameter change. Since errors are small (ppm level) you can ignore the error of the error. Meaning, that I did not calculate, that the change in opamps voltage will change the zener current which will change the output voltage. Because it will be in the 0.001 ppm level, irrelevant.
So the only parameter you need is 0.3 (or 30%) for the gain resistors 0.0001 = 0 for the current resistor.
That is certainly true. And les drifty. And with top shelf components, it is possible to make a better reference than from a IC.
« Reply #24 on: June 08, 2017, 01:12:43 am »
Thanks for pointing that out NANDBlog, that's a simpler way to think about it.
I've written a script which spits out a bunch of combinations of resistors, along with what Vz they are suited for, how much current would run through Rf and Rg, etc.
This script will allow you to find a set of resistors tailored for each individual LM399.
This github gist has the script along with the generated output tables for the Stackpole RNMF series of resistors ($0.10, 50ppm/C) and Yageo MFP ($0.46, 25ppm/C):
https://gist.github.com/cellularmitosis/eef3564c692acc4b590869112042e0f6
(I can't control the order of the files, so you'll have to scroll past the python script to see the data tables).
I've also attached the script to this post.
An example of using the tables:
I have run 1mA through an LM399 and measured a Vz of 6.965V.  I go with the next highest value available in the table (6.9677V, which will yield an Iz of just over 1mA) and look at my options.  If I wanted to have the option of running from a single 9V battery, I'd go with the 12R, 62R, 36K combo, which only requires the opamp to swing up to 6.9797V.  If instead I wanted to minimize the current through Rf+Rg, I might go with the 27R, 620R, 160K combo (44uA), etc.
