Hi Frank,
I'll do that, but I'm not hopeful as with decade A set to CAL then S1A should be open, S1B should connect to R1001, and S1C to R301/R1005.
So R302 and R311 should not be shunted by anything at all.
I do wish the schematic had the colour coding of the wires as well ... I know they follow the usual conventions, but ...
I disconnected the black wires from position one of the A decade switch which should isolate R311 from the switch. Sadly I still get a reading of 9.794k\$\Omega\$ between connectors 3-4 on the oil bath.
To save time I cut the white wire leading to position 21 on the oil bath and measured R302. Sadly it still reads 9.710k\$\Omega\$
I shudder to think how many Mars Bars Fluke would demand to fix this.
Thanks
Dave