I have 4 34470DMMS, 3 are calibrated from Transcat , but when I apply the 0 ohm shot provided by keysight, they all read either negative resistance or 0.01XX ohm range.
I want to use it to calibrate a HP 5.5 digit meter, but from the calibration instruction for the 2 wire, I need to apply short to the terminal in 2 wire mode, but different kind of short yield different "0". I dont have a good short at hand so I was using a Keysight 1mohm shunt module but after calibration, with a ponoma banana plug box with silver wire short, I was reading 0.11ohm. (4 wire calibration I can do with my Fluke resistance standard and all ranges are spot on..)
I understand no one will use 2 wire mode to measure low resistance but that inconstancy in 0 ohm short reading is bugging me...
Also I noticed that 34410A and 34401A reads far better and stable with my own short (they all out of cal for at least 6 years)