I've just used REW software and an EMU0404 soundcard for signal generator and spectrum analyser assessment of some aspects of the 5065A, with signal injection to the A7 AC amp J1 input (in lieu of the optical cell) via 10M and ground clip.
A7 AC Amp module J1 input stage to TP2 shows low harmonics for 0.5Vpp in, with +5.4dB gain. TP1 shows 0.14% H2 with +0.7dB gain from input, and clearly shows the 274Hz notch in the noise floor response. Y output clips for signal levels above ~0.19Vpp at J1, and non-clipped output has low HDs.
As Corby indicates, the 6010 version of A7 presents a 2nd H meter level even for 137Hz signal, and meter level varies with 234 and/or 137Hz magnitude.
With 137Hz signal input, the following A8 Phase detector module generates an Error meter reading that cycles between -50 and +50, and the reference 137Hz oscillator shows sync with external REW signal at 137.035Hz so is ok.
Switching from Open Loop to Operate mode shows the Control meter reading swings and the Integrator Limit indicator lights at <-50, and >+50, and the A9 Integrator output to OCXO EFC- control line cycles from -14.4V to +7.8Vdc, and my SD6250A frequency counter just recognises a change in the 5MHz frequency (which is below the 1Hz resolution of the meter, but identified by toggling of the last digit).
I was also able to confirm the 1MHz and 100kHz outputs are ok, and the A3 multiplier module has 5.4Vdc on TP2, so that aspect is ok.
The Lamp Oven meter swing has subsided a bit to between 15 and 38, so moving in the right direction.
Ciao, Tim