Hi all,
I was curious about how a handheld DMM is designed and how input division and protection was implemented so I bought a fluke 107 just to take it apart and trace back everything. (since it is cheap and very simple)
When I tried to measure the input impedance, I got about 11MOHM from 34401A, 34470A and DMM7510, but when measure with 3478A I got about close to 18Mohm.
When I connect the 1M and 10M resistance standard to 3478A the read out is spot on.
Is it possible that during the resistance measurement mode, the output voltage/current is smaller than other DMMs so it reads way off from other DMMS? (or it is too high and tipped some active disconnect ?)
I am going to cal the 3478A in the 2 wire mode since I received the calibration short for it. but I heard the 30MOHM range cannot be calibrated on some DMM, is that true ?
also, I noticed there is a piece of "Glass" looking plate on the PCB , and I can see finger print on it, not sure what the purpose of that is for....
Image credit from Z
Thank you
Regards,
Li