To expand on that, TAR is calculated as:
$$TAR=\frac{\textrm{Tolerance of device under test}}{\textrm{Accuracy of calibration source}}$$
So to calculate it you need the specified tolerance of your DMM at whatever calibration interval it's calibrated to (generally 1 year unless the customer requested otherwise) and the accuracy of the calibrator that's being used. If you know they use a Fluke 5700A that is calibrated every 90 days, you could look up the specs and get an idea what the accuracy might be.