Thanks for all.
I will pursue a double mixer / interval timer as a long-term project.
Although a DMTD is the gold standard for a time-nut measurement system, you should consider an interim solution. A TAPR-TICC (
https://tapr.org/product/tapr-ticc/ ) is a manageable cost device that will give you an ADev noise floor of 6e-11 @ 1 sec. You can make a lot of good measurements at that level. The graph I posted earlier was made with a Fluke PM-6681 which has a noise floor of about 5e-11 @ 1 sec. The TAPR-TICC can be used later with a DMTD front-end to get to crazy low numbers.
There are also counters like the SR-620, HP 5370A, or perhaps the Wavecrest DTS-20xx that can get to levels of 2e-11 or 3e-11. They can be pricey and all the typical caveats apply when buying used boat anchors.
Just to be clear, I'm not suggesting you buy the TAPR-TICC and later get one of the counters. Choose one or the other as a stepping stone to a DMTD system.
just a reply to your question
This is a 1 GS/s two channel digital scope. It cost all of $200 new.
Why are you being so shy? You're claiming that your Rb-disciplined standard is off-frequency. You've discussed your procedure - which is non-standard. After being asked twice, I still don't know what scope you're using. That's not the way it's done Ken. Particularly when you're making a claim this serious, you should provide as much info as possible regarding the equipment you're measuring, the equipment you're using to make the measurements, and the methodology you're using. Then others can look at your info and see if they can find any problems with your work. You've done some of this, but not all.
the scope is sync'd to channel 1 which is the 10 MHz sinewave of the Z3801A. It puts the zero crossing of that waveform dead-still in the center of the screen at 0.00 ns.
then I watch the channel 2 trace (which is the DUT) and manually move an on-screen cursor to the zero crossing point. I start my stopwatch and record the zero crossing point some minutes later. The scope reads out the time in nanoseconds and it has two cursors. It is quite easy to measure the phase difference to within 0.1 ns. with this method. I just have to watch it so that it does not go beyond 100 ns (slipping a full cycle) because then the results are ambiguous - did it slip one cycle, two?. The unit I am testing takes about 43 minutes to slip a cycle. Like I said - this only works if the DUT is drifting in one direction steadily. And that is the situation I have.
I can potentially resolve 0.1 ns in 2400 seconds: that is 4E-14. Almost a nutty-level.
Anything in the e-14 range is most definitely a nutty-level. However, considering what you're measuring and how you're measuring it, it's likely at the too-good-to-be-true level.
I also have a cheap 1 GSPS digital scope - a Hantek DSO5202B. It specs the delta time measurement error as: "1 sample interval + 100 ppm x reading +0.6 ns". TBH, I don't know how to translate that spec into an error value for your measurement. In any case, we should be using the spec for your scope. But running the measurement for 43 minutes makes me think the error is going to build to something horrible.
My scope also specs frequency measurement as 6 digits of resolution and a *typical*, not worst case, error of 30 ppm. As you can see, it can't be considered to be a precision measurment device. Low end scopes never are.
I asked the seller if there is anything I can adjust on the GNSS RB but he has no clue - If it is phase-locked to the GPS signal, in principle there should be nothing to do. Frequency-locked might be another story I suppose.
If it's not in the manual, there's not much you can do. The X72 is a smart box with both hardware and software frequency adjustments, but I don't know if your unit has a way to access either of them.
Ed