My point is very simple. I repair and test 8.5d meters and 7.5d calibrators, so as part of troubleshooting I always have to do INL test on DUT, to ensure functionality built on top of INL performance are correct. Using full transfer spec as a limiter in such tests is not adequate. You should be able to understand why already. Testing INL shows dodgy out of spec 3458 ADC very easily, even though meter
still meets transfer specifications, thanks to ACAL.
There would be no reason in testing any calibration issues or ACAL or transfer features if core ADC or DAC cannot deliver the required sub-ppm INL performance. Since my tests are also automated, I use two or more verified 3458A's as "composite 0.1ppm 10V INL standard" in such tests, with two DCV meters confirmed lack of the temperature dependency. This is why INL number in datasheet is important for me. Don't have my own JVS yet. Other alternative is manual labor-intensive testing with 720A, SR10*0 and alike. As your Fluke linked doc outlines, INL is also not a problem for transfer between the same nominal value standards, typically used in metrology.
and not all units may be the same.
Precisely! I'd say 2 meters are never same, actually.
Manufacturer specifications apply for most of the units out there. But each particular unit still behaves little differently. More listed design parameters in documentation help me to separate expectations, as I obviously can also perform measurements inside the DMM to find out each domain error and compare that to expected manufacturer values.
With an ADC like in the 3458 there is a chance that a drifting resistor array may also effect the INL
My measurement results (from >8 different ADCs, good and bad) show that all drifty/bad ADCs cannot meet 0.1ppm INL in -11 to +11V sweep. Good ADCs have no problem with this.
There is no absolute need for INL specs. Transfer specs for enough cases would be sufficient. However these are usually only give for a few conditions and not even the same for all meters. Not sure if this is more like cherry picking or just manufacturer specific preferences.
It is more of a desire item, then actual metrology field need. We are deep down the rabbit hole for this one here.
Maybe I'm expecting too much, thinking about a customer who paid $20k+ for "reference DMM" and not having that little bit of optional information? Kind of same goes to traceabillity in calibration reports. Desired wish - to see data of all upstream standards chain to SI, so one can calculate own numbers for U. Like open-source calibrations, sorta. Actual metrology field tho provide you only final measurement result with assigned uncertainty and expecting you to trust the lab in their measurement. Lab is audited by accredited body to ensure their measurements are in order, but there is no way to be 100% sure your particular calibration report is correct. Maybe the calibration tech had bad mood during your DMM calibration
So it's all marketing and specmanship, one would say, no any different to Vishay's "typical 0.0 ppm/C TCR" resistors? Truth is hiding somewhere in between...