Hi, I 've got a wt1600 repaired, changed the sample resistance  in the sample board, now I test It and find the error is out of tolerance. I have got Fluke calibrator, but I don't know how to get in the calibration mode of wt1600, can some one provide a service manual of tell me how to get in the calibration mode, many thanks.
  • Country: cn
Re: YOKOGAWA WT1600 Power meter calibration mode
« Reply #1 on: January 03, 2022, 09:17:00 pm »
there is a adjustment procedure for wt230, maybe I can try this method.
  • Country: cn
« Reply #2 on: January 07, 2022, 02:21:40 am »
That doesn't work on a wt1600, any other suggestions?
Offline PE1RQF
« Reply #3 on: January 14, 2022, 03:09:22 pm »
Try switch on instrument while pressing display button. After start up, press shift and than display button.
Regards,
Erik
  • Country: cn
« Reply #4 on: January 26, 2022, 06:44:57 am »
It's really great, that works. Really thanks.
I will try to calibrate it.
  • Country: cn
« Reply #5 on: January 26, 2022, 08:45:13 am »
 you are supper great.
Offline PE1RQF
« Reply #6 on: January 26, 2022, 05:53:22 pm »
Fine that it is working
regards,
Erik
  • Country: cn
« Reply #7 on: January 27, 2022, 06:28:23 am »
One more thing  I 'm not quite sure. There are 100V(500)  100V(5.5k) 1A(500) 1A(5.5k), does 500 5.5k means the frequence of the standard voltage or current, and for the other values, apply DC signal would be ok?
Offline dl1640
  • Country: cn
« Reply #8 on: January 27, 2022, 12:11:39 pm »
Apply rated signal at 60Hz
  • Country: cn
« Reply #9 on: January 30, 2022, 02:20:17 pm »
Thanks, I will try that and report then.
Offline PE1RQF
« Reply #10 on: January 30, 2022, 02:53:01 pm »
Apply rated signal at 60Hz is correct. it is the filter adjustment.
  • Country: cn
« Reply #11 on: February 03, 2022, 03:08:26 pm »
now I'm a little confused. As in the picture I uploaded, for
10V,apply 10V 60Hz
15V, apply 15V 60Hz
30V,apply 30V 60Hz
60V apply 60V 60Hz
100V apply 100V  60Hz
150V apply 150V 60Hz
100V(500) apply 100V 500Hz
100V (5.5k) apply 100V 5.5kHz
Are these right?
Offline dl1640
  • Country: cn
« Reply #12 on: February 04, 2022, 12:38:08 pm »
That's correct.
(500Hz), (5.5kHz) for measurement under LPF set to ON.
The reference signal applied is at commercial frequency 60Hz/50Hz will be fine.
  • Country: cn
« Reply #13 on: February 04, 2022, 01:55:20 pm »