Dave my point was that Calibration and Verification are NOT interchangeable like you use them. Calibration is a process, and verification is a procedure. All you are saying is that by default the lab does verification, and doesn't tune the UUT, which is understandable. I don't disagree with that, but stating that calibration is ONLY verification is false.
3: to standardize (as a measuring instrument) by determining the deviation from a standard so as to ascertain the proper correction factors
4: to adjust precisely for a particular function
Because of your popularity you have a lot of power to greatly misinform public, remember that.
Sorry for bumping this old thread but i wanted to give my two cents.
I used to work in an ISO17025 accredited cal lab, according to our governing standards body we used the following definitions (roughly, see below for the VIM definitions):
* Calibration: Establish a relationship between a physical variable and an indication (for meters) or another physical variable (for transducers).
* Verification: Use calibration data to determine if the meter or transducer performs within the customer's specifications
* Adjustment: Alter the behavior of the meter or transducer, usually with the intention of making it conform to it's original or new specifications
So, when a customer brought a meter to get calibrated with adjustment we did the following sequence of operations: Calibration, Verification. If the verification failed we did Adjustment, Calibration and Verification again. If the second verification failed again we would contact the customer to decide what to do with the instrument. In some cases the instrument could be repaired, calibrated and verified, in others it would just get junked and replaced.
Some instruments and artifacts can not be adjusted, for example RTD transducers. In those cases we would only do Calibration, or Calibration and Verification, if the verification failed it usually meant the device was damaged and was thus replaced. We were legally required to preserve the Calibration history of the instruments since they were used for measuring the flow of natural gas at points of transfer ($$$$$$). If an instrument had been measuring high or low for a period of time between calibrations both gas companies would be extremely interested to know about it
.
By definition, calibration is ALSO adjustment:
http://www.merriam-webster.com/dictionary/calibrate
Merriam-Webster is not an authority in metrology last I checked, when talking about metrology you should really use the definitions in the VIM (International Metrology Vocabulary) as published by the BIPM.
http://jcgm.bipm.org/vim/en/index.html2.39 calibration:
operation that, under specified conditions, in a first step, establishes a relation between the quantity values with measurement uncertainties provided by measurement standards and corresponding indications with associated measurement uncertainties and, in a second step, uses this information to establish a relation for obtaining a measurement result from an indication
2.44 verification
provision of objective evidence that a given item fulfils specified requirements
3.11 adjustment of a measuring system
set of operations carried out on a measuring system so that it provides prescribed indications corresponding to given values of a quantity to be measured
As it's usual with standards bodies, the above definitions are wrapped in even more jargon but are very precise, and roughly equivalent to the ones i worded at the beginning of the post.