Hello Dr. Frank,
There are 11(out of 13) chips were marked "??" in Chinese which means dead/halted in the photo I posted. This is the most serious error that U180 may cause to an 3458A. The halt may be occasional or only on DC range(according to the hand writing on the chips).
There is one marked as "AC measurement Ok", which probably means DC will be halted as well.
There is only one saying "workable but large tempco", I assume there is no error message associated with it.
I've just asked the one who provided the photo(thy888), he said they were all halted with error message 114, and the auxiliary message being either Multi-slope ADC error or Time Out error.
There are two root causes, one is the severe mismatch of resistors, the other is the bad input switching JFETs of the ADC. There are two paths for the signal, one is the high resolution one(50k resistor) the other is the high speed one(10k resistor).
I once try to fixed an 3458A with large daily drift but failed, there is no error message, just like mine. At the end, I suspect it was due to the U180, had it replaced, and the meter is Ok. There was a thread at 38hot that record my work:
http://bbs.38hot.net/thread-436-1-1.html http://bbs.38hot.net/thread-4104-1-1.htmlFor that error 202(Slave Test: Convergence), it is caused most probably by open JEFT(cannot be closed) of the high resolution switch for 50k resistor, but the high speed JFET switch will probably work, this can be verified by setting the NPLC to a very small value(so that only high speed path is used) and see if the meter will work. And yes, those 10k/50k resistors and JFET switches are all part of the U180 chip that cannot be replace alone easily.
Edit: add a pic from HPJ 1998-04