With thanks to TiN and I believe SvanGool for putting design up on OshPark, and Lou et al for the Aus Cal Club
A set of 6+ LTZ references being slowly built- aim is long term - kept ON - in a controlled temp cabinet, measured every 3 months or so - watch their drift (and that of my 3458 and 34461).
Difficult part will be getting lab at a constant temp without too much interference from heating-cooling system- I don't have a cellar - but tempted to dig a big pit!
I will probably heat / cool lab for 48hrs prior to measure time - then switch off. I have also seen a little interference from the computer +/- PSU / GPIB interface. Lots of known (and avoidable) interference from LED lights and other SWPSUs.
Most (5) will be with TE UPS50 resistors, closest I can easily get to 13k is 13.3k (paralleled 20k and 40k), 70k (20k+50k), I have one set of Mr Pettis' left.
Four will be with LT1006 and 100nF across 1k and 120R, but two LT2057s
Four will be with LTZ1000As but a generous Aussie gave an LTZ1000C and I have one 'C' of my own.
It won't be a quick result project!
Robert
VK5RC