Author Topic: DMM Noise comparison testing project  (Read 230184 times)
0 Members and 11 Guests are viewing this topic.
Offline _pv
  • Country: hn
Re: DMM Noise comparison testing project
« Reply #600 on: June 15, 2022, 02:41:31 pm »
10hours noise & stability of 33years old V2-38 nanovoltmeter:
10uV range no filter no dumping. Plotted raw data.
could you please share raw data as well?
Offline bsw_m
  • Country: by
« Reply #601 on: June 15, 2022, 07:03:50 pm »
No problem!
Measurements were made on the range of 10 μV
First column is time in unixtime format. Second column - data from nanovoltmeter.
« Last Edit: June 15, 2022, 07:05:58 pm by bsw_m »
The following users thanked this post: _pv
Offline alanambrose
« Reply #602 on: June 16, 2022, 11:05:24 am »
Ah, two conversations going at once :)
Re Kleinstein's thoughts:
>>>  So it is not only the 1 PLC case and some of the more disappointing case are included.
As the only no-filter specs were 1 nplc, those were the only ones I used for comparison with my no-filter tests. I agree that the specs are confusing - maybe innocently, they're using some 'real world' examples. And I don't pretend to fully understand the noise specs because of the '2D' combination of nplc & filter data points. There's probably some rationale there but I don't quite see what it is. Is the response time really 'the time you realistically need to wait for the results to settle' rather than an input variable and should therefore really be one of the rightmost columns?
For the 34420A (below) I noted the official specs on the original graph for the 1mV range of 16 and 1.4 ppm at 1 & 200  nplc - although right now (the actual tests were done some time ago), I can't see where I got those from. However, I've also added the 100 nplc official specs on the graph (also below). All said, I'm thinking my K2182A & 34420A results roughly match the official specs and so my analysis is good. Any thoughts on that?
Alan
p.s. I still need to double check the two 2182A auto zero settings ('Front Autozero' and 'Autozero') I used in the test.
« Last Edit: June 16, 2022, 11:57:08 am by alanambrose »
“A foolish consistency is the hobgoblin of little minds"
Online Kleinstein
« Reply #603 on: June 16, 2022, 12:03:44 pm »
The digital filter is simple averaging. So one has to multiply the PLC setting and fitler setting to get the total integration time. Many meters automatic convert PLC settings larger than 10 to real intgration of some 10 PLC and than use averaging as digital filtering.
Offline _pv
  • Country: hn
« Reply #604 on: June 16, 2022, 02:04:40 pm »
Just updated the plot,
and "noise" here is StDev over 10 consecutive measurements at given averaging time
« Last Edit: June 22, 2022, 10:25:14 am by _pv »
The following users thanked this post: bsw_m, zrq
Offline alanambrose
« Reply #605 on: June 16, 2022, 03:06:29 pm »
OK understand, that makes perfect sense.
Alan
Offline n_haku
  • Country: ru
« Reply #606 on: October 13, 2022, 12:53:41 pm »
Mine data from ad7461a. Though cal expired approx half year ago, it should not affect in adc zero noise measue. Data are srtaight, w/o any math, after well warmed, (>2h), rear input shorted with bare copper wire, every run 30 min lenght.
If matters, I can collect data for longer periods, when dmm not used.
PS: as there seems no dedicated thread for equipment's ROMs, I'll also attach dumps for tr6861, r6144, r6551. I checked function work before read.
My lord! We need more precision!
The following users thanked this post: Mickle T., MegaVolt
Offline MegaVolt
  • Country: by
« Reply #607 on: October 13, 2022, 12:57:47 pm »
What is the y-axis dimension?
Offline n_haku
  • Country: ru
« Reply #608 on: October 13, 2022, 01:38:58 pm »
Since no math applied, its in lsb of range, ie 10uV for 10V, 1uV for 1V, 0.1uV for 100mV, and cause max readings are 1e6, it become ppm of range.
The following users thanked this post: MegaVolt
Offline zrq
« Reply #609 on: January 22, 2023, 01:58:07 pm »
Reviving this thread with some data taken from a Keithley 2001M with front input shorted.
It's noise is very disappointing:
1. almost a order of magnitude higher than modern 7.5 digits and some 6.5 digits DMMs
2. can hardly be averaged down at 100s scale (very significant 1/f)
3. 200 mV range is particularly noisy, not really providing higher resolution than the 2 V range in long averaging time.
I'd like to sell the meter and try to get something else (oh no, I don't want to be a voltnut and spend x5 the money for a DMM7510).
Here shows the allan deviation of various different settings, unit in Volt RMS and in ppm range for easier comparison.
The raw data is attached in the compressed files in the next post, first column is time in unix timestamp and second the voltage. A few traces have (low-resolution) room temperature recorded in a separated file. The BATT ones are measuring a 9V battery, not included in the allan deviation plot.
BTW, does anyone know if it's possible to completely disable the screen to save VFD life? The screen off command kept the VFD on with a fixed screen disabled message, increasing the sampling rate at 1PLC from 19.7Hz to 22.2Hz, completely useless.
Edit: added a few traces in the volt scale plot.
« Last Edit: January 28, 2023, 11:11:06 pm by zrq »
Offline zrq
« Reply #610 on: January 22, 2023, 02:08:03 pm »
Selected raw data for Keithley 2001M, a few long and high speed recordings cannot be uploaded due to attachment size limit.
Offline bsw_m
  • Country: by
« Reply #611 on: January 28, 2023, 03:56:44 am »
Noise and stability the R3003 microvoltmeter - part of R3003 voltage comparator.
Microvoltmeter in 10uV range. The output digitized by DMM 100PLC speed.
First column in file - time in UNIX time format. Second column - digitized result which is given to the used range of the microvoltmeter. No more mathematics was used.
The following users thanked this post: Mickle T., Andreas, MegaVolt, Victorman222
Online iMo
  • Country: bt
« Reply #612 on: January 28, 2023, 01:05:27 pm »
Years back I did with my 34401A at 100mV/100NPLC/AZON with 42nV stddev/rms.
I want to repeat the measurement at .1V/1V/10V - what NPLC should I use? 10 or 100?
« Last Edit: January 28, 2023, 01:36:24 pm by imo »
Online Kleinstein
« Reply #613 on: January 28, 2023, 01:40:42 pm »
There is somewhat limited need to have a run at 100 PLC, as this would averaging 10 PLC data internally. So a long enough rund at 10 PLC could be used as well. The Allan deviation plot should take care of the averaging already and the curves should overlap to a large part.
1 PLC may be interesting, but for the 34401 this has quite some noise from quantization. Still possibly a point for the 100 mV range as less 1/f noise is involved.
Online iMo
  • Country: bt
« Reply #614 on: January 28, 2023, 01:44:40 pm »
Ok, I will do die langen Runden with 10PLC (fast 6 digits)..
Offline zrq
« Reply #615 on: January 28, 2023, 02:01:20 pm »
Recently, I observed another surprising behavior of the Keithley 2001M. When the option slot is empty, whether it's covered or not can have a significant effect on the zero point.
My meter comes with neither option card nor the slot cover, so I simply filled the port with some aluminum foil. The observation began with noticing a sudden jump in a long recording of 20V range, followed by another jump back close to the original value later in few thousands of seconds. Initially I believe it's a jump in temperature or something else, but a similar jump happened again when recording for 200V range and this time I'm sure everything is stable. Later, I noticed another similar jump when I was working on the GPIB cable and torn off the filling aluminum foil accidentally, and the value quickly restored after I put another piece of foil filling the option port!
For now, my best explanation of the observation is the internal temperature distribution can be slightly changed by the airflow. The aluminum foil can get loose and move a bit. So far, after I cover the option port with tape and foil more carefully, I haven't seen such jumps ever again.
« Last Edit: January 28, 2023, 02:04:12 pm by zrq »
Online iMo
  • Country: bt
« Reply #616 on: January 28, 2023, 05:29:16 pm »
Here is a 34401A measurement at 10PLC, 100mV range, AzOn, not fully warmed-up, raw data and put into Plotter app. Is this Overlapping Allan Dev the measure we are interested in? Hopefully my settings are ok..
Edit: nope, wrong tau0, fixed..
PS: at 4.04sec tau (100PLC) it shows 38nV which fits my 42nV@100PLC stddev measurement above..
« Last Edit: January 28, 2023, 06:02:42 pm by imo »
Online Kleinstein
« Reply #617 on: January 28, 2023, 06:04:59 pm »
The Allan deviation is a useful form to show the result, as one can also see how good one could get with averaging. Still not everyone is familier with this format.
The simple standard deviation for a fixed window (the lenght 100 or 1000 points is debateable) is easier to understand. So one should give that number too.
The x axis should be in seconds. Normally 10 PLC should be close to 2.5 or 3 readings per second - so the data indicate 50 Hz mains frequency. I think the plot is wrong and need the sampling time and not the samplig rate entered as tau0, as the curve should start at some 0.4 seconds and not 2.5 seconds.
Not fully warmed up adds the drift, that could well be the reason for the upward terend after some 2000 units. So especially the low ranges should be well warmed up.
The following users thanked this post: iMo
Online iMo
  • Country: bt
« Reply #618 on: January 28, 2023, 06:15:19 pm »
The Allan deviation is a useful form to show the result, as one can also see how good one could get with averaging..
Thus my above measurement says I can go for 5minutes with averaging at that range.
I messed with TimeLab a lot when playing with OCXOs, a pity the Plotter app cannot read csv files, I had to replace commas and rename the file..
Online Kleinstein
« Reply #619 on: January 28, 2023, 09:38:25 pm »
The Allan deviation curve again going up is likely due to the drift, not the intrinsic noise. With a a more stable temperature the drift should get smaller and thus even longer averaging may work. There will finally be a limit from temperature fluctuations or background signals (e.g. hum, EMI).  The noise level is already not bad, considering that much of the noise is due to the resistor string (8*13K) used for the protection and the balancing resistor in the zero link.
Online iMo
  • Country: bt
« Reply #620 on: January 29, 2023, 07:45:13 am »
Ok, after warming up my 34401A here are - the 100mV and 10V MADEVs at 10NPLC and AZON over night runs and 1V measurement made over the day.
Needs some analysis on this method and results..
PS:
a) removed the Plotter outputs
b) added 10PLC 100mV, 1V and 10V MADEV in TimeLab
Note: The 1V range has been sampled during the day with more temperature drift around, will be repeated over night sometime..
« Last Edit: January 29, 2023, 01:48:56 pm by imo »
Offline Krisztián
  • Country: hu
« Reply #621 on: July 30, 2024, 04:54:51 pm »
Is there anyone else whose Agilent/Keysight multimeter had the fan installed backwards, and that's why it was noisier (also electrically)?
I bought my 34410A multimeter in 2010. I also have an HP 34401A. I have always wondered why Agilent claims it is an upgrade over the old HP when it is much noisier electrically! It was obvious from the jump in the digits with shorted inputs (I didn't forget Auto Zero, it was on) and also from the histogram I made in 2010. The original NMB-MAT fan has always bothered me. It whined. I put in a Sunon Maglev 2 weeks ago (SUNON MC30101V2-A99) in the same wrong direction as the NMB was (thinking that Agilent knows better, it must be that way for some reason), the whining has stopped but it didn't get much better. Yesterday I got fed up and turned the fan around so it finally blows air out of the multimeter like it should. After that, the shorted 10PLC noise in the 100mV range dropped from 2uV P-P to 1.5uV P-P. Also, the acoustic noise from the multimeter became much more tolerable and much more even.
Someone took 14 years of my life with a fan installed backwards!  :-DMM
« Last Edit: July 30, 2024, 05:48:44 pm by Krisztián »
Offline HighVoltage
« Reply #622 on: July 30, 2024, 06:00:13 pm »
That is an interesting comparison.
In addition to a new fan blowing outwards, I have installed a 100 Ohm resistor in line with the new fan on all my 34410A and 34411A, to keep the speed of the fan down. I think they will last forever now.
The following users thanked this post: Krisztián
Offline Dr. Frank
« Reply #623 on: July 30, 2024, 09:38:28 pm »
Hi, I have the successor of the 34410/411, i.e. the 34465A with the same LM399 reference.
KS definitely made another mechanical design flaw, as the air flows across the reference circuit, creating electrical noise.
I've put a plastic box over the LM399 and aligned circuit, putting down the noise significantly. I have published this effect somewhere, but don't remember the factor at the moment.
It's also possible to replace the LM399 by a standard LTZ1000 circuit, as in the '470, further reducing the noise.
I the 34970A, HP made another mechanical design flaw. The fan is located on the rear of the left side, blowing inside. The fan is half blocked by the mains filter, and therefore creates audible noise. If it would run freely, it would be absolutely silent.
Check if anything is blocking the fan in the 34410A
Frank
« Last Edit: July 30, 2024, 09:41:27 pm by Dr. Frank »
Online dietert1
  • Country: br
    • CADT Homepage
« Reply #624 on: August 01, 2024, 06:34:56 am »
While testing a GPIB to fiber ethernet interface i made, i got a Keithley 2700 log of 1.2 million consecutive 5 PLC readings with shorted front input, line sync and autozero enabled.
Allen variance exhibits a minimum of about 10 to 15 nV when averaging for 300 seconds. Applying a numerical TC compensation will help at longer time scales. The 77xx plugins include temperature sensors.
Regards, Dieter