My Datron 1281 developed an error 2893 which points to the Ohms assembly after I return from vacation :palm: Switching in pathway 056 and found the system giving a value of 0.010078. I tried to do a internal source cal but it end up to the same error 2863. I have swapped the M201 and M202 (DG211CJ) with the same result
Could anybody please share the the value of pathway 056.
Offline doktor pyta
Re: Datron 1281 error 2893 repair
« Reply #1 on: November 22, 2018, 02:40:54 pm »
On my D1281, P056 gives +0.010025V (last digit unstable).
Don't give up :)
« Last Edit: November 22, 2018, 02:42:47 pm by doktor pyta »
  • Country: hk
« Reply #2 on: November 26, 2018, 09:53:45 am »
Dear doktor pyta, thanks for the info.
Did some measurements on the resistor R209 and R210 found within limit. Cleaned the board with IPA yield the same result. Will do a external cal later as suggest by another member.
« Last Edit: November 26, 2018, 12:01:20 pm by syau »
Offline pelule
  • What is business? It’s other people’s money
« Reply #3 on: December 05, 2018, 12:15:16 pm »
I am sorry to post that late, but I was offline due to a surgery some time.
I just read your post.
This error occurs when the internal sources have drifted over time.
You just need to due a "internal source calibration" to solve that problem.
+ Enable calibration (key)
+ go to cal menue
+ select ext cal
+ quit (do not execute)
+ next you will be asked if you want to do the internal source calibration - trigger it
But the 1281 must be powered on at least 48hrs to ensure thermal stabilisation.
Thats the way to continue operating over long periods without calibration.
/PeLuLe
You will learn something new every single day
The following users thanked this post: TiN, ManateeMafia, syau
  • Country: hk
« Reply #4 on: December 06, 2018, 03:38:12 am »
Thanks pelule for the tips  :-+ and the self test now passed.