...Do you have a decent calibrator? ...
Unfortunately, or maybe fortunately, I’m only at the entrance of the metrology rat hole so I don’t have anything that I can use to calibrate the Datron 1081. All I can hope to achieve is to adjust it to correspond to low-grade and uncalibrated standards.
Status update:
1. After leaving the machine unpowered for several days, and the weather turning colder, it booted up in an odd state. Both DC and ohms were unrealistic (negative ohms?). After several minutes of this I turned it off and back on and it was fine. I have not been able to replicate this so far.
2. Because all the ranges are off in the same direction by roughly the same amount I thought I would attempt the “Standard” adjustment that adjusts for drift of the voltage reference. I bought what I had read was the correct calibration key, and it was not. I then tried shorting the key contact and I still can’t get it to shift into “cal” mode. I’ll have to start troubleshooting that one.
3. I’d like to run some long-term logging runs and see how it behaves with respect to drift. However, I’m leery of wearing out the display. On many meters, you can turn the display off via GPIB. I don’t see a way to do that on the 1081. I’m thinking of adding a physical switch to shut down the +175V power supply. There is room at the back of the machine for such a switch.
This is turning into a bit of a slog. Not a Tin-style epic slog, but a slog nevertheless.