Author Topic: Group buy - low thermal EMF LEMO connector for nanovoltmeter (34420/2182)  (Read 6979 times)
Offline HighVoltage
With the new original short, I calibrated the 34420A to 0 OHM and the value shown on the 34420A stays very stable between 0.000,000,0 and 0.000,000,1 even for days and with temperature changes.
When I install my self made LEMO short and let the instrument settle for a few minutes, the value shown is always between 15.9 around 18.6 uOhm, depending on temperature in the lab.
So, the pure copper original connector makes a difference.
But this difference is larger than I had expected.
« Last Edit: June 20, 2020, 12:28:45 pm by HighVoltage »
The following users thanked this post: MegaVolt, syau
Offline CalMachine
  • Metrology Nut
That could potentially be coming from your shorting connection on your DIY plug
All your volts are belong to me
Offline HighVoltage
Agree...
Although I use a pure copper rod, squeezed it between the 4 solder studs and made sure I had a good mechanical connection to all 4 gold plated studs. Then added solder.
I should try a different configuration.
Here is a picture of my gold plated LEMO connector with copper rod soldered to all 4 studs.
Online dietert1
  • Country: br
    • CADT Homepage
Instead of trying to simulate a 4W ohms measurement of a short, you could try to make two independent shorts for drive and sense, maybe with a 1 KOhm resistor between the two shorts. In order to keep the drive current separate. And the original has that plastic sleeve around the "thermal head".
Regards, Dieter
Offline martinr33
If the pins are gold plated, then they are brass not copper. I have both brass and copper pin shorts, so I will have to try measuring both. Maybe you could also check for the voltage measured across the short?
I ran a quick test on a 34420a that I have had running for a couple of weeks using shorts that I made. The Lemo short uses cadmium-tin alloy. These are built as a 4-way short.
Copper pin Lemo connector with nut:   -12nV   0.2nR
Brass pin Lemo clone push/pull             -7nV    0.3nR
I'd trust the Lemo copper pin as the better of the two, but the differences are very small.
The clone shorts are very convenient because of the push/pull connection.
« Last Edit: June 21, 2020, 06:31:21 pm by martinr33 »
