My Hameg 8118 LCR meter only goes up to 200kHz, but it's perfectly adequate for showing what a wirewound resistor looks like.
I'm not familiar with the Hameg 8118, if it supports a sweep mode that is awesome. Unfortunately with my LCR meter I can only switch between a fixed set of measurement frequencies (40Hz、50Hz、60Hz、80Hz、100Hz、120Hz、150Hz、200Hz、250Hz、300Hz、400Hz、500Hz、600Hz、800Hz、1kHz、1.2kHz、1.5kHz、2kHz、2.5kHz、3kHz、4kHz、5kHz、6kHz、7.5kHz、10kHz、12kHz、15kHz、20kHz、25kHz、30kHz、40kHz、50kHz、60kHz、75kHz、100kHz、120kHz、150kHz、200kHz). So basically I can only collect data points but not sweep over an entire range.
So get it "pretty good," as far as materials, and let the temperatures stabilize, and then always assume you're going to need to reverse the excitation current and average the two readings...
So far the resistor doesn't even seem to be that sensitive to ambient temperature change I assume because it is sealed in the metal shell.
That's easy to do on a bridge - my meter "zero" (the midpoint) often slowly shifts around as I flip the generator polarity back and forth. I balance for the midpoint, rather than worrying about zeroing the meter or trying to eliminate the EMF. At balance, the needle doesn't move:
-0 + TEMF == +0 + TEMF.
I think digital meters do that, too.
My meter has an auto zero function (basically you measure a short and open leads). As long as the ambient temperature stays roughly constant and I don't swap the leads zero point calibration works pretty good. Of course If I swap the leads or ambient has changed I need to repeat the process.
As mentioned earlier, you need a sweep of Z against frequency, to which you then fit a model.
I just use an audio soundcard and REW software to do impedance sweeps up to 96kHz.
...
As indicated, multi-section wound parts of what appears to be scatter wound sections are likely to have multiple resonances and may only have a valid simple R-L model up to only lowish frequency, but I guess that depends on what frequency you want to test up to.
Because of trobbins original comment:
I would have thought an impedance versus frequency sweep would have been more enlightening
I came up with the scope/function generator setup as workaround for the missing sweep feature of my LCR meter. Basically since we know R = 3600 Ohm, and we can measure the phase angle for any given frequency with the scope we could load the data into a spread sheet and calculate Z for all datapoints. Unfortunately my scope is not up to the task. First data recording with the software does not work. Also while the scope is good enough to calculate a phase angle at an arbitrary fixed frequency, when performing a sweep there are glitches where the phase angle suddenly jitters significantly even though frequency steadily increases. I recorded a short demo showing the problem
here. This might just be an averaging/on-screen display issue but without working data logging I don't know for sure.
So currently I'm limited to measuring Z/phase angle at any of the frequencies listed above with my LCR meter. However I'd be curious to know if such a measurement setup works better with higher tier scopes.