I am trying to figure out how to correctly use the guard terminal and garud switch for voltage and resistance measurement. But there are some conflicting information in different manuals.
3458a Calibration Manual say on connecting 10K resistance:
"Connect the resistance standard to the multimeter's front panel HI and LO
Input and HI and LO Sense terminals as shown in Figure 3-3. If using a Guard
wire (as shown in Figure 3-2), set the Guard switch to the Open position. If not
using a Guard wire, set the Guard switch to the To LO position."
So they recommend leaving the guard floating. I find that rather odd.
Same for voltage calibration:
"Connect the voltage standard to the multimeter's front panel HI and LO Input
terminals as shown in Figure 3-2. If using a Guard wire (as shown in
Figure 3-2). set the Guard switch to the Open position. If not using a Guard
wire, set the Guard switch to the To LO position."
The 3458a user manual say about guarding:
"The measurement connection illustrations in this chapter show the multimeter's
Guard terminal connected to the low side of the measurement source (guarded
measurements). This configuration provides maximum effective common mode
rejection (ECMR) on the input terminals selected by the Terminals switch,
assuming the Guard switch is in the Open (out) position. For non-guarded
measurements, depress the Guard switch (TO LO position) and do not connect the
Guard terminal to the measurement source. In the TO LO position, the Guard
switch internally connects the Guard terminal to the LO Input terminal on the
terminals selected by the Terminals switch. This configuration provides reduced
ECMR."
The for pictures of measuring voltage and resitance with guard, guard is connected to low on voltage source or resistor and guard switch should be open.
The 742b manual say though:
"Make sure the GUARD terminals of all
interconnected instruments are tied to earth ground at one point and one
point only in the system, and all LO or COM terminals are tied to GUARD
at only one point in the system."
Summary:
3458a calibration manual. Guard to Guard, Guard switch open.
732b manual. Guard to Guard. Connect Guard to ground at one point. Connect Guard to Low at one point (i.e. use the guard to low switch on 3458a, or probably better, short ground, guard and low on 742b side)
I like what the 742b manual says most.... But I would think those who wrote the 3458a calibration manual know what they are writing. And they have had like 35 years to update it if something was wrong....
Anyone have any more thought on this? Any experments made?