Is TimeLab (I am using current, v1.55 Beta) restricted to 200MHz top frequency?
I am trying to use Keysight 53230A to measure 492.52MHz signal and TimeLab is refusing to let me enter anything over 200MHz - either manually or via "Use incoming data to configure measurement"
Is this deliberate feature or a bug?
It seems that FA-2 is allowed to have 492.52MHz but not 53230A.
Cheers
Leo
Offline TheSteve
  • Living the Dream
Re: Is TimeLab restricted to 200MHz measurements on 53230A?
« Reply #1 on: December 13, 2021, 08:28:04 pm »
Are you using the 350 MHz input for a 492.52 MHz signal?
VE7FM
« Reply #2 on: December 13, 2021, 08:41:23 pm »
Yes, is this a problem?
Offline KE5FX
    • KE5FX.COM
« Reply #3 on: December 14, 2021, 10:53:45 pm »
The 53230A code inherited that limit from my 5370B code base, it looks like.  I went ahead and effectively removed the cap by changing it to 100 THz for all of the GPIB counters, just to keep it from being a problem in the future, and posted a slipstream update of 1.55b (date changed, but not the version).  You can reinstall from here.
The following users thanked this post: Leo Bodnar, TiN, ManateeMafia, 2N3055, syau, eplpwr
Offline TheSteve
  • Living the Dream
« Reply #4 on: December 15, 2021, 03:10:39 am »
If it works then it is no problem. I know they added some firmware frequency limits but maybe it was only to the optional 6 and 15 GHz inputs.
VE7FM