Re: Input impedance in Fluke autoranging meters
« Reply #25 on: October 25, 2017, 08:23:02 pm »
Sorry, I'm not allowed to post photos.

The setup isn't very special. A 3458A, a 34470A and a FuG high voltage supply (HCP series with high stability and low noise options). Everything is connected with coaxial high voltage cables and 10kV-SHV Plugs. The HV probes are simply connected to a wire which is soldered to a 10kV-SHV jack.

The DIY 1000:1 divider is built up of 20 selected Caddock USF 371 resistors and a 100k Vishay S102 (or Z201?)+some Rhopoint PWW for fine tunning. Worst case uncertainty calculation gives about 35ppm at 10kV, but it seems to be much better.
 

Offline FlexibleMammothTopic starter

  • Regular Contributor
  • *
  • Posts: 65
  • Country: de
Re: Input impedance in Fluke autoranging meters
« Reply #26 on: October 26, 2017, 01:18:57 pm »
I'm genuinely amazed by the precision of the Testec probes. With fluke, it seems some kind of lottery - you may or may not get lucky... :-BROKE

Thank you so much for the 50mv/500mV measurements. I do not have the means to do these at home, probably need to take my meter to the lab at university. Now I have something to compare to!  :clap:
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5503
  • Country: de
Re: Input impedance in Fluke autoranging meters
« Reply #27 on: October 26, 2017, 01:46:11 pm »

The DIY 1000:1 divider is built up of 20 selected Caddock USF 371 resistors.

Wow... they are 30 Euro a piece !
Is that correct?
There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline FlexibleMammothTopic starter

  • Regular Contributor
  • *
  • Posts: 65
  • Country: de
Re: Input impedance in Fluke autoranging meters
« Reply #28 on: October 26, 2017, 01:49:48 pm »

The DIY 1000:1 divider is built up of 20 selected Caddock USF 371 resistors.

Wow... they are 30 Euro a piece !
Is that correct?

Well as he can't post pictures, this is a cal lab/workplace/whatever. I doubt they care if they can afford both a 3458A and a 34470A   ;D
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5503
  • Country: de
Re: Input impedance in Fluke autoranging meters
« Reply #29 on: October 26, 2017, 03:30:19 pm »

Wow... they are 30 Euro a piece !
Is that correct?

Well as he can't post pictures, this is a cal lab/workplace/whatever. I doubt they care if they can afford both a 3458A and a 34470A   ;D
[/quote]
Well, I also have a 3458A and a 34470A and know about expensive equipment. Somehow I just did not expect this high price for this resistor. But I guess one has to spend the money to get this precision divider.

I have a few Tektronix high voltage probes P6015A and a couple Fluke 80K-40.
It would be nice to have such a precision divider to check the calibration of the probes.

 
There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline FlexibleMammothTopic starter

  • Regular Contributor
  • *
  • Posts: 65
  • Country: de
Re: Input impedance in Fluke autoranging meters
« Reply #30 on: October 26, 2017, 03:56:00 pm »
Well, I also have a 3458A and a 34470A and know about expensive equipment. Somehow I just did not expect this high price for this resistor. But I guess one has to spend the money to get this precision divider.

Mark me down as impressed and envious :-)

It would be nice to have such a precision divider to check the calibration of the probes.

You're not the only one there!
Is 30 bucks for factory-preselected resistors? If not, that setup might have cost even more...
 

Offline fcb

  • Super Contributor
  • ***
  • Posts: 2125
  • Country: gb
  • Test instrument designer/G1YWC
    • Electron Plus
Re: Input impedance in Fluke autoranging meters
« Reply #31 on: October 29, 2017, 04:34:09 pm »
Sorry, I'm not allowed to post photos.

The setup isn't very special. A 3458A, a 34470A and a FuG high voltage supply (HCP series with high stability and low noise options). Everything is connected with coaxial high voltage cables and 10kV-SHV Plugs. The HV probes are simply connected to a wire which is soldered to a 10kV-SHV jack.

The DIY 1000:1 divider is built up of 20 selected Caddock USF 371 resistors and a 100k Vishay S102 (or Z201?)+some Rhopoint PWW for fine tunning. Worst case uncertainty calculation gives about 35ppm at 10kV, but it seems to be much better.
Such a shame about the pix.

I looked up FuG - their gear looks amazing. Very cool that they can build a 100+ person company making that stuff.
« Last Edit: October 29, 2017, 04:37:17 pm by fcb »
https://electron.plus Power Analysers, VI Signature Testers, Voltage References, Picoammeters, Curve Tracers.
 

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5503
  • Country: de
Re: Input impedance in Fluke autoranging meters
« Reply #32 on: October 30, 2017, 12:06:16 pm »
I looked up FuG - their gear looks amazing. Very cool that they can build a 100+ person company making that stuff.

Yes, FUG is an amazingly good company. And when I repaired my 12.5 kV unit, I got all the support I needed, although this instrument is out of production for a long time.
There are 3 kinds of people in this world, those who can count and those who can not.
 

Offline emg

  • Newbie
  • Posts: 4
  • Country: ir
Re: Input impedance in Fluke autoranging meters
« Reply #33 on: February 17, 2024, 05:55:59 am »
thanks for your answer.
My HIOKI clamp meter and UNI-T DDMs seems to use similar divider therefore having variable input resistance. 
 

Offline EC8010

  • Regular Contributor
  • *
  • Posts: 112
  • Country: gb
Re: Input impedance in Fluke autoranging meters
« Reply #34 on: February 17, 2024, 09:26:17 am »
I just checked my Fluke 89 IV direct voltage ranges and got 10.991M, 10.092M, 10.002M, 10.002M.
 


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf