Author Topic: Analog output overvoltage protection  (Read 834 times)
Online ArdWarTopic starter
  • Country: sc
« on: August 19, 2024, 03:30:02 am »
Commercial calibration references are often specced to survive 1000V backfeed into its outputs, but with a caveat of several tens of mA max current. Sounds nice, that's exactly how TVS and steering diode protection work. However I don't think such strategy will survive a low impedance backfeed. At least not if the output also need to provide actual low impedance during normal operation (in addition to the low effective impedance afforded by 4-wire connection).
So, how should I protect from someone hellbent on connecting a power supply to my devices? No need for 1kV cat ratings. I **think** 50V should be enough, lab PSUs over 30V aren't that common.
ADG5401F sounds nice and it also gives me building block for 4-wire connection, but I'd like to hear other's opinions on this.
« Last Edit: August 19, 2024, 04:35:37 am by ArdWar »
Offline Smokey
  • Not An Expert
Re: Analog output overvoltage protection
« Reply #1 on: August 19, 2024, 04:14:20 am »
What is the output impedance of your analog line?  What is it supposed to do?  Is it high bandwidth?  Etc..
Online T3sl4co1l
    • Seven Transistor Labs
« Reply #2 on: August 19, 2024, 04:21:49 am »
Back-to-back depletion MOS might be of interest.  Maybe with a sense divider afterwards, so the op-amp can correct for noise/error of the limiter device.
Bourns TBU things might also be worth a look: https://www.bourns.com/products/circuit-protection/tbu-high-speed-protectors-hsps
Seven Transistor Labs, LLC
Offline Geoff-AU
« Reply #3 on: August 19, 2024, 05:45:10 am »
Place a sidactor hefty enough to blow up the other equipment first ;D
TBUs can be a good choice too if the RDSon is acceptable.  The reference application in the datasheet shows what you need to do if over-voltage protection is also required. SIDAC or gas discharge tube (GDT) are good for OVP.
  • Country: sc
« Reply #4 on: August 19, 2024, 05:49:37 am »
It's a voltage source/reference, no need for high bandwidth. Impedance might be wrong terminology choice here, since it relates more to drive headroom (e.g I can't put a 100R sense resistor if my supply is only 3V above my output and I want to source 30mA).
Duh, for some reason my mind completely blanked on the usual depletion mos/jfet limiter. Got too focused on the usual two BJT current limiting buffer often posted here :palm:
Nonlinearities should be compensated by the 4-wire, so there should be no problem there. Although gotta need to test if it does rob me some precious ppm.
That Bourns device looks nice to try. It does foldback protection!
Didn't expect such thing available from Bourns out of all manufacturers lol.
« Last Edit: August 19, 2024, 05:55:25 am by ArdWar »
Offline Kleinstein
« Reply #5 on: August 19, 2024, 07:21:02 am »
A simple solution can be a PTC type fuse and sufficent clamps / TVS at the actual amplfier. To get a reasonable low output resistance one would likely want a separate path for the feedback, unless there are separate drive and sense terminals anyway. The feedback path can have protection with simple resistors (still with suficient voltage rating).
Back to back depleten FETs are an option, if needed for a lower maximum current with a resistor in between.
Instead of depeten FETs one could also use normal MOSFETs and a PV OK to provide a gate voltage and 2 BJTs and a resistor between the FETs to limit the current to a slight better defined value and less overall resistance.
This version can also work as a switch to turn the output off.
Most solutions would add some resistance and thus a bit extra headroom needed.
