Author Topic: Hameg HZ186 4 Terminal Transformator Test Cable - Internal Circuit  (Read 2126 times)
Offline michael_ngTopic starter
Does anyone have a picture or schematic of the internal circuit of these testadapter for the HM8118 LCR-bridge?
Are there any additional circuitry inside to make valid transformer measurements?
Many thanks in advance.
Offline F64098
Re: Hameg HZ186 4 Terminal Transformator Test Cable - Internal Circuit
« Reply #1 on: October 16, 2018, 08:36:23 pm »
Hello,
when i read all accesible manuals, there is real nothing in this adaptor.
The main thing of this adaptor is the description of the connectors, so that you connect the DUT in the right way.
Best regards
Frank
Offline branadic
  • Sounds like noise
« Reply #2 on: October 16, 2018, 09:12:34 pm »
There is nothing inside, it is just a 4W connection to both terminals. I once disassambled one of them, as I had to develope an application specific test adapter.
-branadic-
Computers exist to solve problems that we wouldn't have without them. AI exists to answer questions, we wouldn't ask without it.